The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The...

The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The instrument was calibrated with a series of standards: 0, 50.00, 100.00, 250.00 mg/L MgSO47H2O. These samples produced average instrument readings of 0.005; 0.192 ; 0.401 ; and 0.993. Triplicate analysis of an unknown yielded instrument readings of 0.328, 0.340, 0.335. What is the average concentration and standard deviation of the unknown? Report your answer as ppm Mg. THE OPTION IS NOT 211

Solutions

Expert Solution

The average concentration is

0.328 + 0.340 + 0.335 = 1.003/3

                                    = 0.334 ppm

Standard deviation
standard deviation =

where, xi is the individual value from population

            n is the number of population

            is the population mean

(xi-)2 = (0.328 - 0.334)2 = (0.006)2 = 0.000036

               (0.340 - 0.334)2 = (0.006)2 = 0.000036

               (0.335 - 0.334)2 = (0.001)2 = 0.00001

      S.D.             = [0.000082/3]1/2

                          = [2.73 x 10-5]1/2

   S.D. = 0.005228 ppm


Related Solutions

Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.041 Absorbance of sample = 0.476 Absorbance of sample plus addition = 0.821 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.044 Absorbance of sample = 0.478 Absorbance of sample plus addition = 0.878 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.043 Absorbance of sample = 0.320 Absorbance of sample plus addition = 0.840 What was the concentration of the cadmium in the waste stream sample? _____ μg/ml...
briefly but Informativey discuss the use of atomic absorption spectrometry in qualitative analysis
briefly but Informativey discuss the use of atomic absorption spectrometry in qualitative analysis
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.034 Absorbance of sample = 0.318 Absorbance of sample plus addition = 0.753 What was the concentration of the cadmium in the waste stream sample? Calculate percent...
In Atomic Absorption Spectrophotometry (AAS) Lab, we ask to make a standard to determine the linear...
In Atomic Absorption Spectrophotometry (AAS) Lab, we ask to make a standard to determine the linear range of the Mg calibration. "First make up a 100.0 ppm Mg solution. Then from the 100 ppm solution make up 25 mLs of 0.5, 1, 2, 5, 10 and 20 ppm Mg standards, then dilute it in a volumetric flask that contain 5 mL of 2% HNO3 solution and the rest with DW" I have to use a micropipette (uL) to prepare the...
In atomic absorption spectrometry, why is "modulation" of the line source necessary for accurate results and...
In atomic absorption spectrometry, why is "modulation" of the line source necessary for accurate results and how is modulation achieved?
Two water samples were analyzed for their copper content by atomic absorption spectroscopy. After 5 replicate...
Two water samples were analyzed for their copper content by atomic absorption spectroscopy. After 5 replicate analyses, the first water sample was found to have 12.6± 0.9 mg/L copper. After 6 replicate analyses, the second water sample was found to have 13.1± 0.6 mg/L copper. Is the copper content in these water samples significantly different? Report your answer in a full sentence, including all necessary information to explain how confident you are about your answer. Answer: Since the calculated t...
How is contribution from atomic emission distinguished from the singal for atomic absorption in atmoic absorption...
How is contribution from atomic emission distinguished from the singal for atomic absorption in atmoic absorption spectroscopy?
What are differences in effects for a chemical and physical interference in an atomic emission spectrometry?...
What are differences in effects for a chemical and physical interference in an atomic emission spectrometry? (Explain??)
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT