Question

In: Chemistry

In atomic absorption spectrometry, why is "modulation" of the line source necessary for accurate results and...

In atomic absorption spectrometry, why is "modulation" of the line source necessary for accurate results and how is modulation achieved?

Solutions

Expert Solution

The function of the chopper is to chop the light leaving the source so that when the incident beam hits the chopper at the solid surface, the beam will be blocked and detector will only read the emitted signal from the flame. As the chopper rotates and the beam emerges to the detector, the detector signal will be the sum of the transmitted signal plus that emitted from the flame. The signal processor will be able to subtract the first signal from the second one, thus excluding the signal from emission in flames. This can be represented by the following equations:

Signal 1 (Blocked Beam) = Pe

Signal 2 (Transmitted Beam) = P + Pe

Overall Difference Signal = (P + Pe) - Pe = P (Corrected Signal)

This correction method for background emission in flames is called Source Modulation.

It should be recognized that addition of extra components to an instrument will decrease the signal to noise ratio and addition of a moving component is usually regarded as a disadvantage due to higher need for maintenance.

Another procedure which can overcome the emission from flames is to use a modulated power supply that will give fluctuating intensities at some frequency (say for example pulsed radiation at a specific frequency). The emission from flames is a continuous signal but that from the source is modulated. Now if we use a high pass RC filter, only the fluctuating signal will be measured as signal while the DC signal will be considered zero as it can not pass through the electronic filter. The high pass RC filter is a device which uses a resistor and a capacitor the impedance of which is inversely proportional to the frequency of the modulated signal. Therefore, only high frequencies will have low impedance and can pass through the capacitor while signals of low frequencies will suffer very high resistance and will not be able to go through the capacitor.


Related Solutions

What is beam modulation and why is it used in atomic absorption?
What is beam modulation and why is it used in atomic absorption?
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.041 Absorbance of sample = 0.476 Absorbance of sample plus addition = 0.821 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.044 Absorbance of sample = 0.478 Absorbance of sample plus addition = 0.878 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.043 Absorbance of sample = 0.320 Absorbance of sample plus addition = 0.840 What was the concentration of the cadmium in the waste stream sample? _____ μg/ml...
briefly but Informativey discuss the use of atomic absorption spectrometry in qualitative analysis
briefly but Informativey discuss the use of atomic absorption spectrometry in qualitative analysis
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.034 Absorbance of sample = 0.318 Absorbance of sample plus addition = 0.753 What was the concentration of the cadmium in the waste stream sample? Calculate percent...
The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The...
The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The instrument was calibrated with a series of standards: 0, 50.00, 100.00, 250.00 mg/L MgSO47H2O. These samples produced average instrument readings of 0.005; 0.192 ; 0.401 ; and 0.993. Triplicate analysis of an unknown yielded instrument readings of 0.328, 0.340, 0.335. What is the average concentration and standard deviation of the unknown? Report your answer as ppm Mg. THE OPTION IS NOT 211
Why would atomic absorption spectroscopy be less sensitive to low source temperature, assuming efficient atomization?
Why would atomic absorption spectroscopy be less sensitive to low source temperature, assuming efficient atomization?
The burner head in the atomic absorption experiment was perfectly aligned with the line of view...
The burner head in the atomic absorption experiment was perfectly aligned with the line of view of the detector (compare to tilting the burner head during the atomic emission experiment). If the burner head was left tilted, how would (1) the sensitivity of your calibration method and (2) the random error in the calculated concentrations of the unknown be affected? (i.e. think what would be affected in your calibration curve)?  
Was the dropping of the Atomic bombs necessary to end the war? Why or why not?
Was the dropping of the Atomic bombs necessary to end the war? Why or why not?