In: Chemistry

In Atomic Absorption Spectrophotometry (AAS) Lab, we ask to make a standard to determine the linear...

In Atomic Absorption Spectrophotometry (AAS) Lab, we ask to make a standard to determine the linear range of the Mg calibration.
"First make up a 100.0 ppm Mg solution. Then from the 100 ppm solution make up 25 mLs of 0.5, 1, 2, 5, 10 and 20 ppm Mg standards, then dilute it in a volumetric flask that contain 5 mL of 2% HNO3 solution and the rest with DW"
I have to use a micropipette (uL) to prepare the standard. How do I prepare the standards of ppm using (uL) micropipette??

Solutions

Expert Solution

we have 100 ppm Mg solution.

from above standared solution we have to prepare 0.5,1,2,5,10 and 20 ppm in 25 ml volumetric flask.

1) 0.5 ppm solution :

C1V1 = C2V2 , FROM dilution principle,

100xX = 0.5X25

X = 12.5 /100 = 0.125 ml of above standared soltion take it in flask and make up with HNO3 and D.W

2)

1 ppm solution :

C1V1 = C2V2 , FROM dilution principle,

100xX = 1X25

X = 25/100 = 0.25 ml of above standared soltion take it in flask and make up with HNO3 and D.W

3)

2 ppm solution :

C1V1 = C2V2 , FROM dilution principle,

100xX = 2X25

X = 50 /100 = 0.5 ml of above standared soltion take it in flask and make up with HNO3 and D.W

4)

5 ppm solution :

C1V1 = C2V2 , FROM dilution principle,

100xX = 5X25

X = 125 /100 = 1.25 ml of above standared soltion take it in flask and make up with HNO3 and D.W

5)

10 ppm solution :

C1V1 = C2V2 , FROM dilution principle,

100xX = 10X25

X = 250 /100 = 2.5 ml of above standared soltion take it in flask and make up with HNO3 and D.W

6)

20 ppm solution :

C1V1 = C2V2 , FROM dilution principle,

100xX = 20X25

X = 500 /100 = 5 ml of above standared soltion take it in flask and make up with HNO3 and D.W


Related Solutions

Standard addition in atomic absorption spectrophotometry: As part of an investigation of Pb in biological ecosystems,...
Standard addition in atomic absorption spectrophotometry: As part of an investigation of Pb in biological ecosystems, some weeds are gathered along a highway where they have been exposed to emissions from gasoline engines. A representative sample of the weeds weighing 6.250 g is ashed to destroy organic matter, and the inorganic residue is dissolved and diluted to 100mL. A portion of this solution is aspirated into an acetylene-air flame, and an absorbance value of 0.125 is found using the 283.31-nm...
The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The...
The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The instrument was calibrated with a series of standards: 0, 50.00, 100.00, 250.00 mg/L MgSO47H2O. These samples produced average instrument readings of 0.005; 0.192 ; 0.401 ; and 0.993. Triplicate analysis of an unknown yielded instrument readings of 0.328, 0.340, 0.335. What is the average concentration and standard deviation of the unknown? Report your answer as ppm Mg. THE OPTION IS NOT 211
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.041 Absorbance of sample = 0.476 Absorbance of sample plus addition = 0.821 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.044 Absorbance of sample = 0.478 Absorbance of sample plus addition = 0.878 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.043 Absorbance of sample = 0.320 Absorbance of sample plus addition = 0.840 What was the concentration of the cadmium in the waste stream sample? _____ μg/ml...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.034 Absorbance of sample = 0.318 Absorbance of sample plus addition = 0.753 What was the concentration of the cadmium in the waste stream sample? Calculate percent...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of the unknown solution. The results obtained are shown in the table. Absorbance Reagent blank 0.028 Sample 0.480 Sample plus addition 0.887 What is the concentration of the cadmium in the waste stream sample? c = μ g /...
Why do we use atomic absorption spectroscopy?
Why do we use atomic absorption spectroscopy?
QUESTION 1 We will be using spectrophotometry to determine the concentration of iron in solutions. First,...
QUESTION 1 We will be using spectrophotometry to determine the concentration of iron in solutions. First, you will generate a calibration curve. The ----- variable is -----, and will be graphed on the x axis of the calibration curve. The -------- variable is ----------, and will be graphed on the --------- axis. Word Bank: -independent -y -absorbance -dependent -concentration QUESTION 2 How many mL of a 100 mg/dL Ferrozine solution are needed to complex 0.00000048 moles of iron? HINT: Look...
Atomic absorption spectroscopy was used to determine the amount of aluminum in a sample of sunscreen....
Atomic absorption spectroscopy was used to determine the amount of aluminum in a sample of sunscreen. A blank solution was measured seven times to determine the detection and quantitation limits of the spectrometer. The blank readings were as follows: 0.0086, 0.0107, 0.0095, 0.0103, 0.0113, 0.0104, and 0.0094 AU. Determine the signal at the detection limit and the signal at the limit of quantitation for this method.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT