Two water samples were analyzed for their copper content by atomic absorption spectroscopy. After 5 replicate...

Two water samples were analyzed for their copper content by atomic absorption spectroscopy. After 5 replicate analyses, the first water sample was found to have 12.6± 0.9 mg/L copper. After 6 replicate analyses, the second water sample was found to have 13.1± 0.6 mg/L copper. Is the copper content in these water samples significantly different? Report your answer in a full sentence, including all necessary information to explain how confident you are about your answer.

Answer: Since the calculated t value is less than the tabulated value (95%, 9 degrees of freedom, tcalc = 0.6, ttable = 2.3) , the two copper concentrations are not statistically different to 95% confidence.

Can you show me how he got to that answer?

Solutions

Expert Solution

t-value calculations :

for Method comparison :

Cu in water sample by AAS:

wo water samples were analyzed for their copper content by atomic absorption spectroscopy. After 5 replicate analyses, the first water sample was found to have 12.6± 0.9 mg/L copper. After 6 replicate analyses, the second water sample was found to have 13.1± 0.6 mg/L copper .

here results, are reported as : mean ± standard deviation

s1 = 0.9 mg/L ; s2 = 0.6 mg/L

first calculate the pooled standard deviation, s :

where s1 and s2 are the standard deviations of the two methods with sample sizes n1 = 5 and n2 = 6.

The standard error of the difference between the two means is calculated as:

The significance level, is calculated using the t-test, with the value t calculated as:

Cu in water : df = n1 + n2 − 2 = 9

s = 0.748; S.E. = 0.453

difference of means = 0.500

t calc (value) city water : 1.103

since calculated t-value is less than value at 95 % confidence level : 2.26

If the absolute value of the t-value is greater than the critical value, you reject the null hypothesis. If the absolute value of the t-value is less than the critical value, you fail to reject the null hypothesis.

the conclusion is that the two means are not significantly different.


Related Solutions

Determination of the Calcium Content of Beer and Water Using Atomic Absorption Spectroscopy 1. Assuming the...
Determination of the Calcium Content of Beer and Water Using Atomic Absorption Spectroscopy 1. Assuming the wort is saturated with Ca3(PO4)2(s), what would happen to the concentration of Ca2+ions in the wort if: (i) More HPO42-ions were added? PO43- Concentration will increase (ii) Ca3(PO4)2 (s) was added? (iii) The pH of the solution was decreased by addition of a suitable acid? 2. Why is LaCl3 utilized? What changes would be needed in set up of the AAS if LaCl3 was...
The purpose of  an ionisation suppressant in atomic absorption spectroscopy is
The purpose of  an ionisation suppressant in atomic absorption spectroscopy is
Three major spectroscopic methods for elemental analysis atoms are Atomic Absorption Spectroscopy, Atomic Emission Spectroscopy, and...
Three major spectroscopic methods for elemental analysis atoms are Atomic Absorption Spectroscopy, Atomic Emission Spectroscopy, and Atomic Mass Spectroscopy. Briefly describe the physics behind each of the three methods.
Why do we use atomic absorption spectroscopy?
Why do we use atomic absorption spectroscopy?
Collimated atomic beam, saturated absorption and two-photon spectroscopy are three methods of Doppler-free laser spectroscopy. (i)...
Collimated atomic beam, saturated absorption and two-photon spectroscopy are three methods of Doppler-free laser spectroscopy. (i) In each case summarize briefly the principle of how the Doppler width is reduced or eliminated , (ii) Which method (methods) collects (collect) signal from all atoms illuminated by laser beams? (iii) In collimated beam spectroscopy the probe laser beam with the wavelength of 0.5 μm intersects the atomic beam (velocity of 300 m/s) at the angle of 70o (the angle between the wave...
1. Samples of drinking water were analyzed for lead contamination. Analysis of the drinking water showed...
1. Samples of drinking water were analyzed for lead contamination. Analysis of the drinking water showed a lead concentration of 12.54 ppb. Analysis of a sample of the same drinking water spiked with 20.00 ppb lead gave a concentration of 29.63 ppb. Calculate the spike recovery. Is the method sufficiently accurate? 2. A chemical technician wants to verify the precision of an analytical procedure used to determine citric acid concentration in orange juice. She obtains 10.00 mL of a particular...
Atomic absorption spectroscopy was used to determine the amount of aluminum in a sample of sunscreen....
Atomic absorption spectroscopy was used to determine the amount of aluminum in a sample of sunscreen. A blank solution was measured seven times to determine the detection and quantitation limits of the spectrometer. The blank readings were as follows: 0.0086, 0.0107, 0.0095, 0.0103, 0.0113, 0.0104, and 0.0094 AU. Determine the signal at the detection limit and the signal at the limit of quantitation for this method.
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of the unknown solution. The results obtained are shown in the table. Absorbance Reagent blank 0.028 Sample 0.480 Sample plus addition 0.887 What is the concentration of the cadmium in the waste stream sample? c = μ g /...
Explain why atomic emission spectroscopy can be used for multi-elemental analysis, but atomic absorption is used...
Explain why atomic emission spectroscopy can be used for multi-elemental analysis, but atomic absorption is used mostly for quantification of single element.
The amount of sodium can be determined in an apple using atomic absorption spectroscopy. 0.4326 g...
The amount of sodium can be determined in an apple using atomic absorption spectroscopy. 0.4326 g of apple was dried, heated at 700oC and chemically treated with a concentrated HCl solution and diluted to 250 mL. 10 mL aliquots were transferred to 50 mL flasks. Each was treated with a standard solution of Na+ at 25.0 ppm and analyzed using flame AAS. Construct a standard addition graph and calculate the weight percentage of Na in the apple. Na+ standard (mL)...
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT