Question

In: Chemistry

Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...

Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.034 Absorbance of sample = 0.318 Absorbance of sample plus addition = 0.753 What was the concentration of the cadmium in the waste stream sample? Calculate percent error caused by using water instead of the reagent blank?

Solutions

Expert Solution

Reagent Blank:

Given the absorbance of reagent blank, Ao = 0.034

Unknown sample:

Corrected absorbance, A1 = 0.318 - 0.034 = 0.284

Suppose the concentration of the unknown sample be, "C1"

Volume of unknown sample, V1 = 10.0 mL

Now A1 = E*b*C1

=> 0.284 = E*b*C1 ----- (1)

Unknown sample + standard sample or spiked sample(10.0 μL of a 1000.0 μg/mL Cd):

Corrected absorbance, A(u+s) = 0.753 - 0.034 = 0.719

concentration of standard, C2 = (1000 μg / mL)

Volume of standard, V2 = 10.0 μL * (10-3 mL / 1μL) = 0.01 mL

amount of unknown sample = C1*V1 = C1 * 10.0 mL

amount of standard = C2*V2 = (1000 μg / mL) * 0.01 mL = 10 μg

total volume of the unknown+standard, Vt = V1+V2 = 10.0 mL + 10.0 μL * (10-3 mL / 1μL) = 10.01 mL

=> Concentration of unknown + standard, C(u+s) = (C1*V1 + C2*V2) / Vt

=> C(u+s) = (C1 * 10.0 mL + 10 μg) / 10.01 mL

Hence A(u+s) = 0.719 = E*b*C(u+s)

=> 0.719 = E*b* [ (C1 * 10.0 mL + 10 μg) / 10.01 mL] ------ (2)

Dividing equaiton-(2) by (1) gives:

0.719/0.284 = E*b* [(C1 * 10.0 mL + 10 μg) / 10.01 mL] / E*b*C1

=> 0.719/0.284 = (C1 * 10.0 mL + 10 μg) / (C1*10.01)

=> 2.5317 = (C1 * 10.0 mL + 10 μg) / (C1*10.01)

=> (C1 * 10.0 mL + 10 μg) = 2.5317 * (C1*10.01)  

=> (C1 * 10.0 mL + 10 μg) = 25.3422*C1

=> 25.3422*C1 - C1 * 10.0 mL = 10 μg

=> C1 = 0.6518  μg/mL

Hence concentration of the cadmium in the waste stream sample = 0.6518  μg/mL (Answer)

When we use water instead of reagent blank, there is no need to subtract 0.034 from absorbance. Hence

0.753/0.318 = (C1 * 10.0 mL + 10 μg) / (C1*10.01)

Solving the above equation gives

C1 = 0.7298

=> Percent error = [(0.7298 - 0.6518) / 0.6518] * 100 = 11.96 % (Answer)


Related Solutions

Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.041 Absorbance of sample = 0.476 Absorbance of sample plus addition = 0.821 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.044 Absorbance of sample = 0.478 Absorbance of sample plus addition = 0.878 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.043 Absorbance of sample = 0.320 Absorbance of sample plus addition = 0.840 What was the concentration of the cadmium in the waste stream sample? _____ μg/ml...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of the unknown solution. The results obtained are shown in the table. Absorbance Reagent blank 0.028 Sample 0.480 Sample plus addition 0.887 What is the concentration of the cadmium in the waste stream sample? c = μ g /...
Copper was determined in a river sample using spectrometry and the method of standard additions. For...
Copper was determined in a river sample using spectrometry and the method of standard additions. For the addition, 100.0 μL of a 1000.0 ppm Cu standard was added to 100.0 mL of river water. Absorbance of river water: 0.572 Absorbance of river water with standard: 1.122 Calculate the concentration of copper in the river water.
a) Horvat and co-workers used atomic absorption spectroscopy to determine the concentration of Hg in coal...
a) Horvat and co-workers used atomic absorption spectroscopy to determine the concentration of Hg in coal fly ash.22 Of particular interest to the authors was developing an appropriate procedure for digesting samples and releasing the Hg for analysis. As part of their study they tested several reagents for digesting samples. Their results using HNO3 and using a 1 + 3 mixture of HNO3 and HCl are shown here. All concentrations are given as ppb Hg sample. HNO3: 161 165 160...
The meethod of standard additions was used to assess the accuracy of the Greiss method for...
The meethod of standard additions was used to assess the accuracy of the Greiss method for the measurement of nitrate. Standard nitrate was addded at 0, 1, 2, 3, and 4 mL to 5 10.00 mL aliquots of sample before being diluted to 25.00 mL in a volumetric flask. If the nitrate standard was 0.96 millimicro and the data is shown below, what is the concentration of nitrate in the original sample? Added Nitrate (mL)    Signal    LINEST   ...
briefly but Informativey discuss the use of atomic absorption spectrometry in qualitative analysis
briefly but Informativey discuss the use of atomic absorption spectrometry in qualitative analysis
The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The...
The magnesium in a series of soil samples was characterized by atomic absorption spectrometry (AAS). The instrument was calibrated with a series of standards: 0, 50.00, 100.00, 250.00 mg/L MgSO47H2O. These samples produced average instrument readings of 0.005; 0.192 ; 0.401 ; and 0.993. Triplicate analysis of an unknown yielded instrument readings of 0.328, 0.340, 0.335. What is the average concentration and standard deviation of the unknown? Report your answer as ppm Mg. THE OPTION IS NOT 211
In atomic absorption spectrometry, why is "modulation" of the line source necessary for accurate results and...
In atomic absorption spectrometry, why is "modulation" of the line source necessary for accurate results and how is modulation achieved?
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT