Question

In: Chemistry

The following data apply to apply to a column for liquid chromatography:                    Length of column:...

The following data apply to apply to a column for liquid chromatography:

                   Length of column: 25.7 cm

                   Flow rate:            0.313 mL/min

                   VM:                      1.37 mL

                   VS:                       0.164 mL

A chromatogram of species A and B provided the following data:

Retention Time, min Width of Peak Base (W), min
Non-retained 3.1 -
A 13.3 1.07
B 14.1 1.16


a)    the expected resolution, if the column length was doubled

b)    the expected capacity factors, if the column length was doubled

c)   the length of column necessary to give a resolution of 1.5

d)   the time required to separate A and B with a resolution of 1.5

Solutions

Expert Solution

a)    the expected resolution, if the column length was doubled

Ans: If the the column length was doubled then the expected resolution also increases by factor 1.44.

Resolution = 2(tRC– tRB)/(WC+ WB)

                  R = 2(14.1-13.3)/(1.16+1.07) = 0.72

As column length is doubled, Resolution is 0.72 X 2 = 1.44

b)    The Capacity Factor () = KB/KA Where, K is a Retention factor

         KA = 13.3 - 3.1 / 3.1 = 3.3, and KB = 14.1 - 3.1 / 3.1 = 3.5

Hence, = 2 X 3.5 / 3.3 = 2.12

c)   The length of column necessary to give a resolution of 1.5

Ans: R1/R2 = (L1/L2)1/2

           0.72/1.5 = (25.7/L2)1/2

Therefore, L2 = 111.5

d)   The time required to separate A and B with a resolution of 1.5

        The time required to separate A from B is 14.1 min for 25.7 cm column . Now, if Column length is increases to          111.5, retention time will also increase

        tR1/tR2= L1/L2

14.1/tR2 = 25.7/111.5 = 61 minutes.


Related Solutions

elect all statements below that apply to Column Chromatography. Column chromatography can be used to purify...
elect all statements below that apply to Column Chromatography. Column chromatography can be used to purify small (< 0.1 g) sample mixtures. The compound in a mixture with the lowest Rf value will exit the column first. Column chromatography is often used to purify large (10 g) sample mixtures. Column chromatography is used mainly as an analytical tool.
Compare and contrast Column (solid-liquid) chromatography, Paper Chromatography, and thin layer chromatography in respect to the...
Compare and contrast Column (solid-liquid) chromatography, Paper Chromatography, and thin layer chromatography in respect to the size of the sample that can be separated, the ease of use, and the versatility of the method.
With respect to liquid column chromatography in which the stationary phase is silica, propose a suitable...
With respect to liquid column chromatography in which the stationary phase is silica, propose a suitable strategy for separating two components of a binary mixture that fail to move when hexane is the mobile phase and move the same distance when methanol is the mobile phase. Explain your reasoning.
Describe how multi-column chromatography can be advantageous over single column chromatography. Include the following terms in...
Describe how multi-column chromatography can be advantageous over single column chromatography. Include the following terms in your explanation: binding capacity, breakthrough
Explain how gas chromatography differs from liquid chromatography.
Explain how gas chromatography differs from liquid chromatography.
Explain and compare thin layer chromatography with column chromatography (in-detailed answer)
Explain and compare thin layer chromatography with column chromatography (in-detailed answer)
When would the application of thin-layer chromatography, adsorption column chromatography, and gas chromatography be appropriate to...
When would the application of thin-layer chromatography, adsorption column chromatography, and gas chromatography be appropriate to use?
During a column chromatography laboratory experiment, a column is to be run with hexane first, then...
During a column chromatography laboratory experiment, a column is to be run with hexane first, then a hexane/ dichloromethane mixture, then 100% dichloromethane. What will happen if you mix up the hexane and dichloromethane? How can you fix the problem?
For each of the following categories of compounds, indicate the type of liquid chromatography that could...
For each of the following categories of compounds, indicate the type of liquid chromatography that could be used to separate analytes in that category. (Drag the appropriate chromatography label to the blank next to the category. Only one label applies to each category, but a label may be used more than once). Seperation Mode Analyte Category Fm>2000,soluble in octane,size10 nm Fm>2000, soluble in butanol, size 100 nm Fm>2000, soluble in water,nonionic,size 10 nm Fm<2000, soluble in in ethanol Fm<2000 ,...
1) Matching. Fill in the number matching column A with the column B _______ adsorption chromatography                      ...
1) Matching. Fill in the number matching column A with the column B _______ adsorption chromatography                          1. Solute equilibrates between mobile phase and surface of stat. phase _______ Reverse phase                                                        2. Effluent from a chromatographic column _______ ion-exchange chromatography                              3. Relates resolving power of a column with parameters that cause peak                                                                                      broadening         _______ molecular exclusion chromatography                      4. The most common type of column used in partition chromatography _______ affinity chromatograhy                                              5. Different size solutes penetrate...
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT