In: Chemistry
Explain why fluorescence is collected at an angle other than 180˚ to the excitation light, and why an emission filter is used.
Fluorescence is mostly observed at a 90° angle relative to
excited light. The angle is maintained 90° other than 180 in
order to avoid interference of the transmitted excitation
light.
There is no such monochromator that is perfect and it will transmit
some stray light, i.e wavelengths other than targeted ones.An ideal
monochromator would only transmit light in the specified range and
While measuring at a 90° angle, only the light scattered by the
sample causes stray light. This results in a better
signal-to-noise ratio, and lowers the detection
limit by approximately a factor 10000 when compared to the 180°
geometry.
The emission filter is typically a bandpass filter that passes only
the wavelengths emitted by the fluorophore and blocks all
undesired light outside this band – especially the
excitation light. By blocking unwanted excitation energy
(including UV and IR) or sample and system autofluorescence,
optical filters ensure the darkest background.