When the patient makes an effort to initiate a breath, and the
ventilator doesn't support it, it is called triggering
asynchrony.
The reasons for this triggering asynchrony are:
Ventilator related cause:
- Malfunction of ventilator sensors
- Inappropriate sensitivity parameter -
- If the ventilator-breath is pressure triggered
- When a patient initiates a breath, there is a negative pressure
developed by the patient's airways.
- This negative pressure is picked up by the ventilator.
- The ventilator then initiates the breath.
- If the patient doesn't initiate enough negative pressure, the
ventilator doesn't initiate a breath.
- If the ventilator-breath is flow triggered
- When the patient initiates a breath, the patient develops a
flow in the airways.
- The ventilator senses the flow and initiates a breath.
- If the patient doesn't initiate enough flow the ventilator
doesn't initiate the breath.
- Prolong inspiratory time - The patient cycles off his breath
before the ventilator cycle off the breath ( breath cycling is
defined as the end of inspiration followed by the initiation of
expiration)
Patient-related causes:
- Respiratory muscle weakness - Chronic illness ( respiratory and
non-respiratory) can lead to muscle weakness. This reduces
respiratory effort.
- Decrease neural drive - ( Due to paralysis of muscle due to
drugs, toxins, neurological disease.)This reduces the neural drive
for respiration and therefore, there is reduced respiratory
effort.
- Dynamic hyperinflation - Due to bronchospasm, there is a
prolongation of expiratory phase of respiration. This leads to
air-trapping.
These problems can be solved in the following ways:
- Change defective sensors of the ventilator.
- Flow triggered breath ( ventilator ) is better than pressure
triggered breath. To initiate a Pressure triggered breaths from the
ventilator, the patient has to develop a significant pressure. Flow
triggered breaths are more patient-friendly.
- Reduce the inspiratory time
- Reduce the level of sedatives and neuromuscular blockade.
- Use bronchodilator to reduce bronchospasm and extrinsic peep,
to reduce the dynamic hyperinflation.