Question

In: Chemistry

Draw the layout of a Michelson interferometer used for FT-IR spectrometry including all elements necessary for...

Draw the layout of a Michelson interferometer used for FT-IR spectrometry including all elements necessary for the generation of interferograms. Describe how the system functions and how spectra are generated.

Solutions

Expert Solution

Michelson interferometer

The Michelson interferometer, which is the core of FTIR spectrometers, is used to split one beam of light into two so that the paths of the two beams are different. Then the Michelson interferometer recombines the two beams and conducts them into the detector where the difference of the intensity of these two beams are measured as a function of the difference of the paths. Figure 3 is a schematic of the Michelson Interferometer  

Figure 3. Schematic of the Michelson interferometer

A typical Michelson interferometer consists of two perpendicular mirrors and a beamsplitter. One of the mirror is a stationary mirror and another one is a movable mirror. The beamsplitter is designed to transmit half of the light and reflect half of the light. Subsequently, the transmitted light and the reflected light strike the stationary mirror and the movable mirror, respectively. When reflected back by the mirrors, two beams of light recombine with each other at the beamsplitter.

If the distances travelled by two beams are the same which means the distances between two mirrors and the beamsplitter are the same, the situation is defined as zero path difference (ZPD). But imagine if the movable mirror moves away from the beamsplitter, the light beam which strikes the movable mirror will travel a longer distance than the light beam which strikes the stationary mirror. The distance which the movable mirror is away from the ZPD is defined as the mirror displacement and is represented by ∆. It is obvious that the extra distance travelled by the light which strikes the movable mirror is 2∆. The extra distance is defined as the optical path difference (OPD) and is represented by delta. Therefore,

δ=2Δ(1)(1)δ=2Δ

It is well established that when OPD is the multiples of the wavelength, constructive interference occurs because crests overlap with crests, troughs with troughs. As a result, a maximum intensity signal is observed by the detector. This situation can be described by the following equation:

δ=nλ(2)(2)δ=nλ

with n = 0,1,2,3...

In contrast, when OPD is the half wavelength or half wavelength add multiples of wavelength, destructive interference occurs because crests overlap with troughs. Consequently, a minimum intensity signal is observed by the detector. This situation can be described by the following equation:

δ=(n+12)λ(3)(3)δ=(n+12)λ

with n = 0,1,2,3...

These two situations are two extreme situations. If the OPD is neither n-fold wavelengths nor (n+1/2)-fold wavelengths, the interference should be between constructive and destructive. So the intensity of the signal should be between maximum and minimum. Since the mirror moves back and forth, the intensity of the signal increases and decreases which gives rise to a cosine wave. The plot is defined as an interferogram. When detecting the radiation of a broad band source rather than a single-wavelength source, a peak at ZPD is found in the interferogram. At the other distance scanned, the signal decays quickly since the mirror moves back and forth. Figure 4(a)shows an interferogram of a broad band source.

Fourier Transform of Interferogram to Spectrum

The interferogram is a function of time and the values outputted by this function of time are said to make up the time domain. The time domain is Fourier transformed to get a frequency domain, which is deconvolved to product a spectrum. Figure 4 shows the Fast Fourier transform from an interferogram of polychromatic light to its spectrum.

(a) (b)

Figure 4. (a) Interferogram of a monochromatic light; (b) its spectrum

The Fourier Transform

The first one who found that a spectrum and its interferogram are related via a Fourier transform was Lord Rayleigh. He made the discover in 1892. But the first one who successfully converted an interferogram to its spectrum was Fellgett who made the accomplishment after more than half a century. Fast Fourier transform method on which the modern FTIR spectrometer based was introduced to the world by Cooley and Turkey in 1965. It has been applied widely to analytical methods such as infrared spectrometry, nuclear magnetic resonance and mass spectrometry due to several prominent advantages which are listed in Table 1.

Table 1. Advantages of Fourier Transform over Continuous-Wave Spectrometry

Fourier transform, named after the French mathematician and physicist Jean Baptiste Joseph Fourier, is a mathematical method to transform a function into a new function. The following equation is a common form of the Fourier transform with unitary normalization constants:

F(ω)=12π−−√∫∞−∞f(t)e−iωtdt(4)(4)F(ω)=12π∫−∞∞f(t)e−iωtdt

in which t is time, i is the square root of -1.

The following equation is another form of the Fourier transform(cosine transform) which applies to real, even functions:

F(ν)=12π−−√∫∞−∞f(t)cos(2πνt)dt(5)(5)F(ν)=12π∫−∞∞f(t)cos⁡(2πνt)dt

The following equation shows how f(t) is related to F(v) via a Fourier transform:

f(t)=12π−−√∫∞−∞F(ν)cos(2πνt)dν(6)(6)f(t)=12π∫−∞∞F(ν)cos⁡(2πνt)dν

An Alternative Explanation of the Fourier Transform in FTIR Spectrometers

The math description of the Fourier transform can be tedious and confusing. An alternative explanation of the Fourier transform in FTIR spectrometers is provided here before we jump into the math description to give you a rough impression which may help you understand the math description.

The interferogram obtained is a plot of the intensity of signal versus OPD. A Fourier transform can be viewed as the inversion of the independent variable of a function. Thus, Fourier transform of the interferogram can be viewed as the inversion of OPD. The unit of OPD is centimeter, so the inversion of OPD has a unit of inverse centimeters, cm-1. Inverse centimeters are also known as wavenumbers. After the Fourier transform, a plot of intensity of signal versus wavenumber is produced. Such a plot is an IR spectrum. Although this explanation is easy to understand, it is not perfectly rigorous.

Simplified Math Description of the Fourier Transform in FTIR

The wave functions of the reflected and transmitted beams may be represented by the general form of:

E1=rtcEm×cos(νt−2πkx)(7)(7)E1=rtcEm×cos⁡(νt−2πkx)

and

E1=rtcEm×cos[νt−2πk(νx+Δd)](8)(8)E1=rtcEm×cos[νt−2πk(νx+Δd)]

Where

• ΔdΔd is the path difference,

• rr = reflectance (amplitude) of the beam splitter,

• tt is the transmittance, and

• cc is the polarization constant.

The resultant wave function of their superposition at the detector is represented as:

E=E1+E2=2(r×t×c×Em)×cos(νt−2πkx)cos(πkΔd)(9)(9)E=E1+E2=2(r×t×c×Em)×cos⁡(νt−2πkx)cos⁡(πkΔd)

Where Em,, ν, and k are the amplitude, frequency and wave number of the IR radiation source.

The intensity (I) detected is the time average of E2 and is written as

I=4r2t2c2E2mcos2(νt−2πkx)cos2(πkΔd)(10)(10)I=4r2t2c2Em2cos2(νt−2πkx)cos2(πkΔd)

Since the time average of the first cosine term is just ½, then

I=2I(k)cos2(πkΔd)(11)(11)I=2I(k)cos2⁡(πkΔd)

and

I(Δd)=I(k)[1+cos(2πkΔd)](12)(12)I(Δd)=I(k)[1+cos⁡(2πkΔd)]

where I(k) is a constant that depends only upon k and I(∆d) is the interferogram.

From I(∆d) we can get I(k) using Fourier transform as follows:

I(Δd)−I(∞)=∫km0I(k)cos(2ΠkΔd)dk(13)(13)I(Δd)−I(∞)=∫0kmI(k)cos⁡(2ΠkΔd)dk

Letting Km →∞, we can write

I(k)=∫∞0[I(Δd)−I(∞)]cos(2ΠkΔd)dΔd(14)(14)I(k)=∫0∞[I(Δd)−I(∞)]cos⁡(2ΠkΔd)dΔd

The physically measured information recorded at the detector produces an interferogram, which provides information about a response change over time within the mirror scan distance. Therefore, the interferogram obtained at the detector is a time domain spectrum. This procedure involves sampling each position, which can take a long time if the signal is small and the number of frequencies being sampled is large.

In terms of ordinary frequency, νν, the Fourier transform of this is given by (angular frequency ω=sπνω=sπν):

f(ν)=∫∞−∞f(t)e−i2Πνtdt(15)(15)f(ν)=∫−∞∞f(t)e−i2Πνtdt

The inverse Fourier transform is given by:

f(ν)=∫∞−∞f(t)e+i2πνtdt(16)(16)f(ν)=∫−∞∞f(t)e+i2πνtdt

The interferogram is transformed into IR absorption spectrum (Figure 5) that is commonly recognizable with absorption intensity or % transmittance plotted against the wavelength or wavenumber. The ratio of radiant power transmitted by the sample (I) relative to the radiant power of incident light on the sample (I0) results in quantity of Transmittance, (T). Absorbance (A) is the logarithm to the base 10 of the reciprocal of the transmittance (T):

A=log101T=−log10T=−log10II0(17)(17)A=log101T=−log10T=−log10II0

Figure 5. IR spectrum of a sample

Hands-on Operation of an FTIR Spectrometer

Step 1: The first step is sample preparation. The standard method to prepare solid sample for FTIR spectrometer is to use KBr. About 2 mg of sample and 200 mg KBr are dried and ground. The particle size should be unified and less than two micrometers. Then, the mixture is squeezed to form transparent pellets which can be measured directly. For liquids with high boiling point or viscous solution, it can be added in between two NaCl pellets. Then the sample is fixed in the cell by skews and measured. For volatile liquid sample, it is dissolved in CS2 or CCl4 to form 10% solution. Then the solution is injected into a liquid cell for measurement. Gas sample needs to be measured in a gas cell with two KBr windows on each side. The gas cell should first be vacuumed. Then the sample can be introduced to the gas cell for measurement.

Step 2: The second step is getting a background spectrum by collecting an interferogram and its subsequent conversion to frequency data by inverse Fourier transform. We obtain the background spectrum because the solvent in which we place our sample will have traces of dissolved gases as well as solvent molecules that contribute information that are not our sample. The background spectrum will contain information about the species of gases and solvent molecules, which may then be subtracted away from our sample spectrum in order to gain information about just the sample. Figure 6 shows an example of an FTIR background spectrum.

Figure 6. Background IR spectrum

The background spectrum also takes into account several other factors related to the instrument performance, which includes information about the source, interferometer, detector, and the contribution of ambient water (note the two irregular groups of lines at about 3600 cm–1 and about 1600 cm–1 in Figure 6) and carbon dioxide (note the doublet at 2360 cm–1 and sharp spike at 667 cm–1 in Figure 6) present in the optical bench.

Step 3: Next, we collect a single-beam spectrum of the sample, which will contain absorption bands from the sample as well as the background (gaseous or solvent).

Step 4: The ratio between the single-beam sample spectrum and the single beam background spectrum gives the spectrum of the sample (Figure 7).

  

Figure 7. Sample IR spectrum

Step 5: Data analysis is done by assigning the observed absorption frequency bands in the sample spectrum to appropriate normal modes of vibrations in the molecules.

Portable FTIR Spectrometers

Despite of the powerfulness of traditional FTIR spectrometers, they are not suitable for real-time monitoring or field use. So various portable FTIR spectrometers have been developed. Below are two examples.

Ahonen et al developed a portable, real-time FTIR spectrometer as a gas analyzer for industrial hygiene use. The instrument consists of an operational keyboard, a control panel, signal and control processing electronics, an interferometer, a heatable sample cell and a detector. All the components were packed into a cart. To minimize the size of the instrument, the resolution of FTIR spectrometer was sacraficed. But it is good enough for the use of industrial hygiene. The correlation coefficient of hygienic effect between the analyzer and adsorption tubes is about 1 mg/m3.

Korb et al developed a portable FTIR spectrometer which only weighs about 12.5 kg so that it can be held by hand. Moreover, the energy source of the instrument is battery so that the mobility is significantly enhanced. Besides, the instrument can function well within the temperature range of 0 to 45 oC and the humidity range of 0 to 100%. Additionally, this instrument resists vibration. It works well in an operating helicopter. Consequently, this instrument is excellent for the analysis of radiation from the surface and atmosphere of the Earth. The instrument is also very stable. After a three-year operation, it did not lose optical alignment. The reduction of size was implemented by a creative design of optical system and accessory components. Two KBr prisms were used to constitute the interferometer cavity. Optical coatings replaced the mirrors and beam splitter in the interferometer. The optical path is shortened with a much more compact packaging of components. A small, low energy consuming interferometer drive was designed. It is also mass balanced to resist vibration. The common He-Ne tube was replaced by a smaller laser diode.

Plz rate it positive


Related Solutions

Give one disadvantage of FT-IR microscopy and electron microscopy. How can FT-IR spectroscopy be used in...
Give one disadvantage of FT-IR microscopy and electron microscopy. How can FT-IR spectroscopy be used in fiber analysis?
1) Draw the energy investment phase of glycolysis including all compounds, enzymes, products used and produced....
1) Draw the energy investment phase of glycolysis including all compounds, enzymes, products used and produced. 2)Draw the energy payoff phase of glycolysis including all compounds, enzymes, products used and produced. 3)List what enzymes can be subjected to feedback inhibition and what they can be inhibited by. 4)Which enzyme(s) catalyzes a reaction that produces ADP during glycolysis? 5)Which enzyme(s) catalyzes a reaction that produces ATP during glycolysis
Draw an axon and synapse with all the necessary components for neurotransmission and then explain (a)...
Draw an axon and synapse with all the necessary components for neurotransmission and then explain (a) how a specific neurotoxin (you find choose one: don’t usew-conotoxin, it’ used below) affects this system and (b) how selective serotonin reuptake inhibitors work on this system and briefly explain how they would be effective as antidepressants.
Explain the necessary requirements for effective market segmentation, including the criteria that should be used when...
Explain the necessary requirements for effective market segmentation, including the criteria that should be used when determining which segments a business should pursue.
Discuss and/or describe how three of these core elements are used in strategic planning:  finance (including efficiency/productivity),...
Discuss and/or describe how three of these core elements are used in strategic planning:  finance (including efficiency/productivity), quality and value, current and planned mix of products and services, current/proposed strategies, and human resource management. Give examples or explain the importance of each.
Question: 1. List ALL the components necessary to build a 100BASE-TX network, including the quantity needed...
Question: 1. List ALL the components necessary to build a 100BASE-TX network, including the quantity needed...
(Show All the Calculations /Wherever Necessary Draw Diagrams) Q 1. An electromagnetic wave with its electric...
(Show All the Calculations /Wherever Necessary Draw Diagrams) Q 1. An electromagnetic wave with its electric field parallel to the plane of incidence is incident from vacuum onto the surface of a perfect conductor at an angle of incidence θ. Obtain an expression for the total electric and the magnetic field.
Point Groups and symmetry operation 1.Explain all the symmetry operations in details 2.Draw the symmetry elements...
Point Groups and symmetry operation 1.Explain all the symmetry operations in details 2.Draw the symmetry elements on the stereogram step by step [1] 222 [2] 2/m [3] 4mm [4] 6(bar)m2 3.List 32 point groups with crystal system.
Hand draw the following transformations step by step, including all intermediates. Ideally, also show electron pushing...
Hand draw the following transformations step by step, including all intermediates. Ideally, also show electron pushing A. Conversion of asparagine to aspartic and isoaspartic acid in peptides. B. Conversion of aspartic acid to isoaspartic acid in peptides. C. Cleavage of the peptide backbone at the aspartic acid residue. D. Discuss whether acid may catalyze each reaction (A, B or C); i.e., whether the reaction is faster at (pH lower than 7) than at pH 7. E. Discuss whether base may...
Promotion: explain all the promotion mix elements currently being used to promote the product. Need to...
Promotion: explain all the promotion mix elements currently being used to promote the product. Need to justify the reason different promotion mix elements chosen (ACT mouthwash line)
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT