Question

In: Chemistry

You are required to prepare standard standards for a calibration curve. All the glassware and pipettes...

You are required to prepare standard standards for a calibration curve. All the glassware and pipettes your method requires have a tolerance (RSD) of 0.05%. You start working solution of 0.2 M analyte. Remove 10.0 mL of that solution with a pipette, add it to a 100 mL volumetric flask and dilute it to the mark. Remove 25.0 mL of the newly prepared (diluted) standard, add to a 500.0 mL volumetric flask, and dilute it to the mark. Report the concentration of standard in the 500 mL flask with uncertainty.

Solutions

Expert Solution

All the glasswares used in the experiment have a RSD of 0.05%. Report the absolute uncertainity and the relative uncertainty as below.

Glassware

Volume Delivered (mL)

Absolute Error (RSD*Volume Delivered) (mL)

Relative Error

10 mL pipette

10.00

(0.005)/(10.00) = 0.0005

100 mL volumetric flask

100.00

(0.05)/(100.00) = 0.0005

25 mL pipette

25.00

(0.0125)/(25.00) = 0.0005

500 mL volumetric flask

500.00

(0.25)/(500.00) = 0.0005

We take 10.00 mL of 0.2 M analyte and dilute to 100.00 mL in a volumetric flask; the dilution factor is (100.00 mL)/(10.00 mL) = 10.00. The concentration of the analyte is the 100 mL volumetric flask is (0.2 M)/(10.00) = 0.02 M.

25.00 mL of the prepared analyte is taken and diluted further to 500.00 mL; the dilution factor is (500.00 mL)/(25.00 mL) = 20.00. The concentration of the analyte in the final solution is (0.02 M)/(20.00) = 0.001 M.

Since we have a RSD of 0.0005 overall, the absolute error in the measured concentration is (0.0005)*(0.001 M) = 5.0*10-7 M.

The concentration of the final solution is reported as 1.0*10-3 5.0*10-7 M (ans).


Related Solutions

Calibration Method 1: Regular Calibration Starting with solid CaCO3, describe how you will prepare Ca2+ standards...
Calibration Method 1: Regular Calibration Starting with solid CaCO3, describe how you will prepare Ca2+ standards in 0.1 M nitric acid in the concentration range 5-60 ppm (MW CaCO3 = 100.08 and MW Ca = 40.08). Remember you do not want to weigh out very small amounts because small fluctuations in the balance can create large errors. You should not weigh out an amount of CaCO3 less than 0.1 g. Therefore, you may need to prepare a stock solution with...
You prepare a mixture to use to create a colorimetry calibration curve for FeSCN2+ by mixing...
You prepare a mixture to use to create a colorimetry calibration curve for FeSCN2+ by mixing these things in a 25.00 mL volumetric flask and then diluting to the volume of the flask. 10.00 mL 0.100 M KSCN 5.00 mL 2 M HNO3 4.00 mL 0.00100 M Fe(NO3)3 What is the concentration of FeSCN2+ in this mixture?
Why is it important that all absorbance readings for the calibration curve be measured at the...
Why is it important that all absorbance readings for the calibration curve be measured at the same wavelength?
Would a change in temperature during a calibration of volumetric glassware experiment (involving a buret and...
Would a change in temperature during a calibration of volumetric glassware experiment (involving a buret and pipet) impact accuracy, precision or both? Explain. What would be the difference if a graduated cylinder was used instead? Explain.
One of the most important standards of the IFRS is standard #15, you are required to...
One of the most important standards of the IFRS is standard #15, you are required to read about it, then to write an essay about it. Explain it in details.
What would be the effect on the slope of the calibration curve if you were to...
What would be the effect on the slope of the calibration curve if you were to choose a wavelength to do the analysis that was 50 nm longer than the wavelength selected in the experiment?
Assuming that you are limited to the standard glassware and equipment available in the Chemistry lab...
Assuming that you are limited to the standard glassware and equipment available in the Chemistry lab room, write a procedure to determine the density of an oil.
Based on your previous experience on the AA, what tends to happen to the calibration curve when more concentrated standards are used?
Based on your previous experience on the AA, what tends to happen to the calibration curve when more concentrated standards are used? Do you expect this to happen for this experiment with AE and (ICP-OES) ? Why or why not?in AA when more concentrated standards are used, the absorbance increased.Atomic Absorption (AA),Atomic Emission (AE),and Inductively Coupled Plasma-Optical Emission Spectroscopy (ICP-OES)
3.  (15 pts) For the following calibration techniques:  (a) Standard addition, (b) External calibration, and (c) Internal calibration:  Draw...
3.  (15 pts) For the following calibration techniques:  (a) Standard addition, (b) External calibration, and (c) Internal calibration:  Draw a picture of what each graph would look like, describe each technique, and give a reason why each might be used for quantification (an advantage of each). *just wants a general example plot, not data is provided for this problem.*
in GC calibration curve, which concentration remains constant and which one changes? (analyte vs internal standard)
in GC calibration curve, which concentration remains constant and which one changes? (analyte vs internal standard)