A technologist is performing a colorimetric assay using 96?well ELISA microtiter plates. The ELISA plate reader measures the absorbance of the solution in each well (cuvet) by recording the light passing through the solution from the top to the bottom of each well. The assay’s color, when tested in a standard spectrophotometer, is linear to an absorbance of 4.0 however the ELISA reader has a range of linearity which only extends to an absorbance of 1.2. When a microtiter plate is placed in the ELISA reader, the technologist notes that the solution in several wells has an approximate absorbance of 1.5. The assay material is very expensive and each assay takes one day to perform. Considerable effort is made to get as much data as possible from each run. Therefore, in order to measure the absorbance of the solution in these wells, the technologist dilutes the samples by adding a volume of buffer equal to the volume in the well. To her surprise, the absorbance values remain the same.
1. Considering Beer’s law, why didn’t the absorbance value change?
2. Are there any steps that can be taken to bring the absorbance within the linear range?
3. Calculate the concentration of the compound using the values below. Show all calculations
Dilution 1:2
Abs std 0.45
Abs unk 0.78
Conc std 4 g/dL
Further you have to also ensure about the following factors of the sample and molecules. There are some limitations of the Beer-Lambert law which are as follow: