In: Chemistry
In an experiment, why would the concentration of a sample decrease when changing the wavelength for a spectrophotometer from 235 nm to 280 nm?
If I understand your question correctly, then you want to ask that why the absorbance (not the concentration) of a sample decreases when we change the wavelength from 235 nm to 280 nm.
This change in absorbance is attributed to several factors which are given below:
1. The absorbance of a particular solution depends upon the identity of compound present in it. It means substance A has different absorbance pattern than substance B.
2. The absorbance of a substance covers a range of wavelength
known as absorption envelope that gradually increases from starting
wavelength reaches its maximum value at a certain wavelength (
known as
max ) of that substance and then decreases again. It
means if the absorbance of compound is measured at
max it will be maximum and lesser on other
wavelengths.
3. Since the concentration of a solution is related to
absorbance by the formula
; where A is absorbance,
is molar extinction coefficient and l is pathlength of cell, so we
can calculate the concentration by recording absorbance at a
particular wavelength provided
at that wavelength is known.
Now coming to your question;
If you are observing a decreased absorbance at 280 nm then it is
because of the fact that wavelength 280 nm is far from the
max of that compound. The higher absorbance at 235 nm is
obtained because it should be near to
max. But this entirely depends on the range of
absorption envelope of that substance and not merely true for
absorbance of all compounds at 235 nm and 280 nm. If a substance
has
max at 280 nm then you will gate higher absorbance at
280 nm.
Lastly the concentration of the solution doesn't change on changing the wavelength. It is the absorption property of that substance which changes with change in wavelength and results in different absorption value.