Question

In: Chemistry

Why is the “standard additions” a better method compared to the external calibration method?

Why is the “standard additions” a better method compared to the external calibration method?

Solutions

Expert Solution

The method is usually presented as the separate addition of several different equally-spaced amounts of analyte to separate aliquots of test solution. Measurement is followed by extrapolation of the calibration line to zero response.The use of several spiking concentrations is justified in the standard paradigm by the idea that it helps to check that the calibration is truly linear. However, this rationale is not compelling in a routine, quality-assured laboratory, because:

• you should not attempt standard additions unless you are quite convinced at the validation stage that the analytical calibration is truly linear over the whole relevant concentration range (nonlinear extrapolation being an unwise enterprise);

• tests for non-linearity require a large number of measurements to have a useful degree of statistical power. That would require far more work for the analyst than is reasonable for obtaining a single measurement result.

Standard addition method:

1.Most convenient when a small number of samples are to be analyzed.

2.Useful when the analyte is present in a complicated matrix and no ideal blank is available.

-Add one or more increments of a standard solution to sample aliquots of the same size. Each mixture is then diluted to the same volume.

-Prepare a plot of Analytical Signal versus:

a)volume of standard solution added, or

b)concentration of analyte added.

The method of standard additions is usually followed to eliminate matrix effects. Experimentally, equal volumes of the sample solution are taken, all but one are separately ‘spiked’ with known and different amounts of the analyte, and all are then diluted to the same volume. The instrument signals are then determined for all these solutions and the results plotted. As usual, the signal is plotted on the y-axis; in this case the x-axis is graduated in terms of the amounts of analyte added (either as an absolute weight or as a concentration). The (unweighted) regression line is calculated in the normal way, but space is provided for it to be extrapolated to the point on the x-axis at which y = 0. This negative intercept on the x-axis corresponds to the amount of the analyte in the test sample. This value is given by a/b, the ratio of the intercept and the slope of the regression line.


Related Solutions

3.  (15 pts) For the following calibration techniques:  (a) Standard addition, (b) External calibration, and (c) Internal calibration:  Draw...
3.  (15 pts) For the following calibration techniques:  (a) Standard addition, (b) External calibration, and (c) Internal calibration:  Draw a picture of what each graph would look like, describe each technique, and give a reason why each might be used for quantification (an advantage of each). *just wants a general example plot, not data is provided for this problem.*
The meethod of standard additions was used to assess the accuracy of the Greiss method for...
The meethod of standard additions was used to assess the accuracy of the Greiss method for the measurement of nitrate. Standard nitrate was addded at 0, 1, 2, 3, and 4 mL to 5 10.00 mL aliquots of sample before being diluted to 25.00 mL in a volumetric flask. If the nitrate standard was 0.96 millimicro and the data is shown below, what is the concentration of nitrate in the original sample? Added Nitrate (mL)    Signal    LINEST   ...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.041 Absorbance of sample = 0.476 Absorbance of sample plus addition = 0.821 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.044 Absorbance of sample = 0.478 Absorbance of sample plus addition = 0.878 What was the concentration of the cadmium in the waste stream sample?Later, the analyst...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.043 Absorbance of sample = 0.320 Absorbance of sample plus addition = 0.840 What was the concentration of the cadmium in the waste stream sample? _____ μg/ml...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of...
Atomic absorption spectrometry was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of solution. The following data were obtained: Absorbance of reagent blank = 0.034 Absorbance of sample = 0.318 Absorbance of sample plus addition = 0.753 What was the concentration of the cadmium in the waste stream sample? Calculate percent...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of...
Atomic absorption spectroscopy was used with the method of standard additions to determine the concentration of cadmium in a sample of an industrial waste stream. For the addition, 10.0 μL of a 1000.0 μg/mL Cd standard was added to 10.0 mL of the unknown solution. The results obtained are shown in the table. Absorbance Reagent blank 0.028 Sample 0.480 Sample plus addition 0.887 What is the concentration of the cadmium in the waste stream sample? c = μ g /...
Copper was determined in a river sample using spectrometry and the method of standard additions. For...
Copper was determined in a river sample using spectrometry and the method of standard additions. For the addition, 100.0 μL of a 1000.0 ppm Cu standard was added to 100.0 mL of river water. Absorbance of river water: 0.572 Absorbance of river water with standard: 1.122 Calculate the concentration of copper in the river water.
Describe the physical effects and interferences that the method of standard additions would correct or compensate...
Describe the physical effects and interferences that the method of standard additions would correct or compensate for in AAS. this question came straight from the textbook... no other explanation was given
21 What major advantage does a standard addition method offer over a calibration graph for the...
21 What major advantage does a standard addition method offer over a calibration graph for the quantitative analysis of a sample?