Variations in source geometry
- When the activity is drawn into a syringe, the source geometry
will be different from that in a vial.
- The composition of the container, thickness, and distribution
will affect the measurement.
- The self-absorption of the emitted radiation will change as the
source volume changes.
Source position
- The manufacturer’s source holder is designed to keep the source
in the area of maximum response on the vertical axis of the
well.
- Variations in response due to changes in vertical height or
horizontal position of a few millimeters are usually
insignificant
Source adsorption
- Certain radiopharmaceuticals have been observed to adsorb to
the surface of the container.
- The adsorbed activity can be a significant percentage of the
total.
- The possibility of activity adsorption should be considered
whenever the facility uses syringes from a different
manufacturer.
The major advantage of this technique is that it directly tests
the accuracy of the dose calibrator for measuring clinical
radionuclides without requiring individual and expensive calibrated
sources for the measurements. Although we did not observe
radionuclide dependence of a calibration source, we prefer to use
133Ba since it emits low and high energy photons has a long
half-life and detects all changes in dose calibrator chamber
absorption characteristics over long periods of time. The procedure
takes minimal time and no source investment is required since any
source currently used for dose calibrator accuracy testing can be
employed.