In: Chemistry
hi,
1. Tetramethylsilane is the accepted internal standard for calibrating chemical shift for 1H, 13C and 29Si NMR spectroscopy in organic solvents (where TMS is soluble). In water, where it is not soluble, sodium salts of DSS, 2,2-dimethyl-2-silapentane-5-sulfonate, are used instead. Because of its high volatility, TMS can easily be evaporated, which is convenient for recovery of samples analyzed by NMR spectroscopy.
2.Solid particles distort the magnetic field homogeneity because the magnetic susceptibility of a particle is different from that of the solution. A sample containing suspended particles thus has a field homogeneity distortion around every single particle. This causes broad lines and indistinct spectra that cannot be corrected. So that there are no solid particles in your samples, you must filter ALL samples into the nmr tube. You should filter samples through a small plug of glass wool tightly packed into a Pasteur pipette. If the plug is not tight enough, the filtration will be ineffective; if it is too big, some of your sample will remain trapped in it. Do not use cotton wool, since most NMR solvents dissolve material from it which can easily be seen in 1H spectra. After filtration the sample should be as clear as water though, of course, not necessarily colourless.
3. Wiping clean the nmr tubes is necessary because it may contain both dust and sample residue as possible contaminants.Failure to wipe off grease and other chemicals off of the outside of the NMR tube will contaminate the spinner and may even cause the spinner to fail to grip the tube properly. So the nmr tube should be always wiped with kim wipes before introducing into spinner.
4. The NMR magnet is arguably the most important part of the NMR spectrometer. The NMR magnet is one of the most expensive components of the nuclear magnetic resonance spectrometer system. Most NMR magnets today are shielded magnets. These magnets have an additional superconducting coil outside of the main coil which cancel out much of the fringe field from the main coil. As a consequence, the stray field outside the magnet is very small. Having a small fringe field becomes important in higher field magnets where safety concerns become more important. This is the reason for spining the tube inside of the magnet
5. Shimming is adjusting the resolution of the signal by
optimizing the homogeneity of the magnetic field. The lines in NMR
spectra are very narrow-linewidths of 1 Hz or less are not
uncommon-so the magnetic field has to be very homogeneous. What we
have to do is to surround the sample with a set of shim coils, each
of which produces a tiny magnetic field with a particular spatial
profile which can be used to cancel out the inhomogeneities in the
main magnetic field. The current through each of these coils is
adjusted until the magnetic filed has the required homogeneity,
something we can easily assess by recording the spectrum of a
sample which has a sharp line.
The shims are labeled according to the filed profiles they
generate. So, for example, there are usually shims labeled x, y and
z, which generate magnetic fields varying in the corresponding
directions.
The axial coils, z1-z4 modify the field profile along the axis of
the main field, which is vertical. The radial coils are x and y,
and other combinations containing one of the two. The radial
inhomogeneities can be averaged out to some extent by spinning.
Adjustment of the axial shims should be done with the sample
spinning. The radial shims are done without spinning. There are
several ways of shimming to optimize the homogeneity of the
magnetic field (shimming on the FID, shimming on the spectrum,
autoshim, gradient shimming and shimming on the lock). The shimming
on the lock is more preferable for a standard user and you will
open the lock window, which allows shimming to be performed on the
lock. When shimming on the lock, you monitor the intensity of the
lock signal. You need to adjust only z1 and z2 shims. Each shim
setting (z1-z4) controls the current through shim coils that
control magnetic field gradients in different directions. It is
important to know that the z direction is parallel to the vertical
direction of the probe and it is for this reason that the height (5
cm) of the sample in the NMR tube affects the z shim settings
rather dramatically.
Hope this was helpful.