A) You are given a dye with an excitation maximum of 358 nm and an emission...

A) You are given a dye with an excitation maximum of 358 nm and an emission maximum of 461 nm, and your fluorescence microscope has fluorescence cubes with the filters below. Do you think you will you be able to see any fluorescence from this dye using any of these fluorescence cubes? Explain your answer. B) Does it make a difference if the barrier filter is a band pass versus a long pass? Why?

Filter cube Excitation (exciter) Emission (barrier) Dichromatic Mirror

WU -> MWU2 330-385 420 400

WB -> MWB2 460-490 520 500

WG -> MWG2 510-550 590 700

Solutions

Expert Solution

A) I think we shall be able to see the fluroscence from the given dye whose excitation is taking place at 358nm and emission is 461nm by using the first dichromatic mirror filter WU-> MWU2 330-385 420 400. The fluroscence filters are strategically oriented within a specialized cube or block that enables the illumination to enter from one side and pass to and from the specimen in defined directions along the microscope optical axis. Now we are showing the transmission spectra as follows in figure 1 and in figure 2

On looking at these figures we can clearly see that the transmission spectra corresponds to the given excitation and emission wavelengths. Fig 1 shows that the filter is Dichoric and 400LP so it meets the specifications for the first option.

B) Yes it will make a difference as the long pass filter allows the transmission for longer wavelengths while bandpass allows the transmission for a certain band of wavelengths. The given wavelengths correspond to blue colour regime. If we want to eliminate the noise that is not spectrally overlapping with the signal then the bandpass filter is a good choice. On the other hand, when the noise spectra is overlapping with the signal than long pass filter is a good option to go with.


Related Solutions

Herb wants to know the wavelength of maximum excitation andemissionto the nearest 0.1 nm for a...
Herb wants to know the wavelength of maximum excitation andemissionto the nearest 0.1 nm for a number of dyes and also wants to get the lowest possible detection limit for each dye (i.e. he wants to maximize signal).  You want to build the most sensitive spectrometer possible for a moderate cost. How would you design a system to meet Herb’s needs.  Be specific about the components.
A. Determine the maximum absorbance wavelength for the dye (Amaranth) in nm (use internet or ask...
A. Determine the maximum absorbance wavelength for the dye (Amaranth) in nm (use internet or ask me for alternative method if you don't have regular access) and in a few sentences explain what is happening when inside the instrument when an absorbance spectra is being acquired. What are the units on the y-axis (ordinate) of the graph? What would be the problem with using a acrylic cell when the sample absorbs in the UV region of the EM spectrum? Explain...
How do you determine if peaks on an emission and excitation spectra are for Raman scattering...
How do you determine if peaks on an emission and excitation spectra are for Raman scattering or Rayleigh scattering? please give example and reasoning
Identify the peaks for Raman or Rayleigh on a emission/excitation spectrum for dilute quinine and for...
Identify the peaks for Raman or Rayleigh on a emission/excitation spectrum for dilute quinine and for dilute sulfuric acid. spectra are not provided
a gas discharge lamp with prominent emission lines at 245nm, 436 nm and 546 nm is...
a gas discharge lamp with prominent emission lines at 245nm, 436 nm and 546 nm is used to illuminate a sample of iron (wavefunction 4.5 eV). Determine the allowed energies for emitted photoelctrons
a. what is the difference between a flourescence excitation spectrum and a fluorescence emission spectrum?. which...
a. what is the difference between a flourescence excitation spectrum and a fluorescence emission spectrum?. which resembles an absorption spectrum. b. consider a molecule that can fluoresce from the s1 state and phosphoresce from the t1state. which is emitted at longer wavelength?make a sketch showing absorption fluorescence and phosphorescence on a single spectrum.
Why is the emission frequency lower than the excitation? What happens to the difference in energy?
Why is the emission frequency lower than the excitation? What happens to the difference in energy?
The emission wavelength lies between 275 nm and 450 nm and the light yield per neutron...
The emission wavelength lies between 275 nm and 450 nm and the light yield per neutron is reported to be 20000 photons/MeV neutron energy (a) What largest work function of a material used for a photo cathode would be acceptable that still allows detection of an emission originating from the scintillator? (b) What is the maximum kinetic energy of electrons emitted from the photo cathode, given the emission wavelength range of the scintillator? (c) How much charge is being generated...
In a hydrogen atom what is the wavelength in nm of the emission line corresponding to...
In a hydrogen atom what is the wavelength in nm of the emission line corresponding to the transition from n=4 to 3
Substance H has maximum absorbance at 550 nm; Substance M has maximum absorbance at 650 nm....
Substance H has maximum absorbance at 550 nm; Substance M has maximum absorbance at 650 nm. A standard solution of H has a concentration 74.2 ppm and an absorbance of 0.972 at 500 nm and an absorbance of 0.276 at 650 nm. A standard solution of M has a concentration of 35.9 ppm and an absorbance of 0.102 at 500 nm and of 0.564 at 650 nm. A sample solution that is a mixture of H and M has an...
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT