In: Chemistry
In Preparative Gas Chromatography with a mixture, 3-pentanol was found to have a starting collection time of 3.81 minutes and an ending collection time of 5.38 min. What is its retention time? Which compounds have higher/lower retention times, and what are some factors that affect retention time?
Retention time is the time taken by the component to pass through the chromatography column. It is the time from the injection of sample into the chromatographic column to the elution of the component. It is usually measured in seconds or minutes.
Each component have different retention time based on interaction with stationary phase inside the chromatography column.
Retention time for 3-Pentanol= (5.38minutes) - (3.81 minutes)= 1.57 minutes
Factors that affect retention time: