Question

In: Chemistry

When an ion-selective electrode for X was immersed in 0.0760 M XCl, the measured potential was...

When an ion-selective electrode for X was immersed in 0.0760 M XCl, the measured potential was 0.0450 V. What is the concentration of X when the potential is 0.0600 V? Assume that the electrode follows the Nernst equation, the temperature is at 25°C, and that the activity coefficient of X is 1.

[X+] = ??M

Solutions

Expert Solution

Voltage of an ion selective electrode can be given using Nerst equation as follows:

E = Eo + (2.303 RT/nF) log (X)

Where ,

E = Potential of electrode ( 0.0450 V)

Eo = Potential of reference electrode ( we have to find out this first)

R = universal gas consatant (8.314 J mol-1 K-1)

T = Absolute temperature (25 + 273 = 298 K )

n = Charge on ionic speceis ( Since species is X+ , it is 1 in this case )

F = Faradys constant ( 96500 Coulomb / mol )

X = Concentration of ionic species ( Since activity coefficient is given as 1 concentration of ion will be 0.076 M )

After plugging all values we get

0.0450 V = E0 + 0.0591 log (0.0760 )

E0 = 0.0450 V - ( - 0.0661 V) = 0.1111 V

Now since we know the potential of reference electrode, we can calculate the concentration of X when the potential is 0.0600 V

E = Eo + (2.303 RT/nF) log (X)

0.0600 V = 0.1111 V + 0.0591 log A

log X = - 0.0511 / .0591 = - 0.8646

By taking antilog we get the concentration of [X+] = 0.1365 M.


Related Solutions

Fluoride Selective Ion Electrode. Describe how a fluoride selective ion electrode measures the fluoride ion concentration...
Fluoride Selective Ion Electrode. Describe how a fluoride selective ion electrode measures the fluoride ion concentration and explain why the potential approaches 0 at high concentrations
Explain the working principle of an ion-selective electrode.
Explain the working principle of an ion-selective electrode.
how does the flouride-ion selective electrode detect the flouride ion?
how does the flouride-ion selective electrode detect the flouride ion?
A solid-state fluoride ion-selective electrode responds to F-but not to HF. It also responds to hydroxide ion at high concentration when
A solid-state fluoride ion-selective electrode responds to F-but not to HF. It also responds to hydroxide ion at high concentration when. Suppose that such an electrode gave a potential of +100 mV (versus S.C.E.) in 10-5 M NaF and +41 mV in 10-4 M NaF. Sketch qualitatively how the potential would vary if the electrode were immersed in 10-5 M NaF and the pH ranged from 1 to 13.
Describe how a fluoride selective ion electrode measures the fluoride ion concentration and explain why the...
Describe how a fluoride selective ion electrode measures the fluoride ion concentration and explain why the potential approaches 0 at high concentrations
How does the Fluoride Ion-Selective Electrode it detect the fluoride ion? Include the species of fluoride...
How does the Fluoride Ion-Selective Electrode it detect the fluoride ion? Include the species of fluoride that need to be considered, how to free the fluoride ion, and pH considerations.
ION-SELECTIVE ELECTRODE POTENTIOMETRY The fluoride ion concentration in a mouthwash sample was determined using ISE. Potentials...
ION-SELECTIVE ELECTRODE POTENTIOMETRY The fluoride ion concentration in a mouthwash sample was determined using ISE. Potentials of standard solutions containing fluoride ion are given in Table 1. The sample for ISE was prepared by pipetting 10.00 mL of the original mouthwash and diluting to 250.00 mL. The potential of the sample was recorded as –0.015 V. Table 1. Measurement of ISE potential of standard solutions containing fluoride ion. [F-], M Potential, mV 5.0 x 10-6 0.077 5.0 x 10-5 0.020...
Lab Purpose: Determine the concentration of fluoride in mouthwash or toothpaste using an ion selective electrode...
Lab Purpose: Determine the concentration of fluoride in mouthwash or toothpaste using an ion selective electrode (ISE) will be used to determine the concentration of F- in a sample of mouthwash or toothpaste. Demonstrate Nernstian electrode response. 1. What is the most common type of ion selective electrode? 2. Why do reactions that remove ions from solution interfere with the operation of an ISE? 3. Why should the data for the calibration plot and unknown values be obtained on the...
145&146&147. The potential of a silver electrode is measured relative to an Ag-AgCl electrode for the...
145&146&147. The potential of a silver electrode is measured relative to an Ag-AgCl electrode for the titration of 100.0 mL of 0.100 M Cl with 0.100 M Ag+. What is the potential after 75.00 mL of titrant is added. Eo = 0.799 V for Ag+, E = 0.197 V for the Ag-AgCl electrode and Ksp = 1.8 × 10−10. A) 0.493 V B) 1.070 V C) 0.521 V
Describe the response mechanism of the fluoride-ion-selective electrode. Explain why the OH- is the major interfering...
Describe the response mechanism of the fluoride-ion-selective electrode. Explain why the OH- is the major interfering ion in F- ISE measrements.