Question

In: Chemistry

A visible light spectrophotometric determination of penicillin makes use of color-forming reaction of the penicillin. If...

A visible light spectrophotometric determination of penicillin makes use of color-forming reaction of the penicillin. If this compound absorbs at 475 nm with an ε = 1560 M-1cm-1, what would be calculated as the limit of detection with an instrument, using a 1.00 cm cuvette, with its smallest absorbance measurement at 0.005?

Solutions

Expert Solution

Ans. Beer-Lambert’s Law, A =e C L                 - equation 1,               where,

                                    A = Absorbance

                                    e = molar absorptivity at specified wavelength

                                    L = path length (in cm)

                                    C = Molar concentration of the solute

Given,

                e = 1560 M-1cm-1

            L = 1.0 cm

                A = 0.005              - smallest absorbance, limiting detection

Putting the values in equation 1

                0.005 = 1560 M-1cm-1 x C x 1.0 cm

            Or, C = (0.005) / (1560 M-1) = 3.20 x 10-6 M

That the as small as 3.20 x 10-6 M concentation of penicillin can be detected using the specified VIS-spectrophotometer.

Thus, detection limit = 3.20 x 10-6 M = 3.20 x 10-6 moles /L   


Related Solutions

In visible light, which color has the longest wavelength?' Are Microwave, Visible Light, Gamma rays in...
In visible light, which color has the longest wavelength?' Are Microwave, Visible Light, Gamma rays in order of lowest to highest energy? are Radio broadcast, cell phone and GPS examples of technologies that uses radio waves
Cell Biology: Microscopes 1) If you could use any color in the spectrum of visible light...
Cell Biology: Microscopes 1) If you could use any color in the spectrum of visible light to observe your specimens in the microscope, what color would you use to obtain the highest resolution? Explain. 2) Nmae 3 differences between upright and inverted compound microscopes. 3) In what circumstances would you use an upright, an inverted and a fluorence microscope? 4) Would you use the same filter cube for every fluorophore you use in your experiment?
M 12: Why can we not use visible light (and optical microscopy) to study nanoscale materials?...
M 12: Why can we not use visible light (and optical microscopy) to study nanoscale materials? What instruments are better suited and why? Explain your answer.
An early method of measuring the speed of light makes use of a rotating slotted wheel....
An early method of measuring the speed of light makes use of a rotating slotted wheel. A beam of light passes through a slot at the outside edge of the wheel, as in the figure, travels to a distant mirror, and returns to the wheel just in time to pass through the next slot in the wheel. One such slotted wheel has a radius of 8.8 cm and 100 slots at its edge. Measurements taken when the mirror is L...
Green plants use light from the Sun to drive photosynthesis. Photosynthesis is a chemical reaction in...
Green plants use light from the Sun to drive photosynthesis. Photosynthesis is a chemical reaction in which water H2O and carbon dioxide CO2 chemically react to form the simple sugar glucose C6H12O6 and oxygen gas O2 . What mass of carbon dioxide is consumed by the reaction of 5.4g of water? Round your answer to 2 significant digits.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT