Question

In: Chemistry

a. The pH meter was improperly calibrated, so that the pH readings on the instrument were...

a. The pH meter was improperly calibrated, so that the pH readings on the instrument were 0.3 pH units too low.

A. pKa would be too low and Ka would be too high

B. pKa would be too high and Ka would be too low

C. neither would change

b. An endpoint/equivalence point was erroneously read as 27.20 mL whereas the reading was actually 26.80 mL.

A. pKa would be too low and Ka would be too high

B. pKa would be too high and Ka would be too low

C. neither would change

D. none of these

c. In Step 2, instead of adding 75 mL of water, 150 mL of water was added.

A. pKa would be too low and Ka would be too high

B. pKa would be too high and Ka would be too low

C. neither would change

D. none of these

Please explain why you chose the answer you chose, thank you!

Solutions

Expert Solution

a) effect of Ka with a pH is marking 0.3 lower...

example:

pH = pKa at half equivalence point...

if real pKa was 4.0, then pH = 4.0

Bu since it is marking0.3 lower, then at the same volume, you will expect a pH = 4-0.3 = 3.7

so the pKa = 3.7, which is false, since it is LOWER

meaning that

Ka = 10^-pKa =10^-3.7 will be HIGHER

choose a

b)

Actual V = 26.8 mL

V false = 27.20 mL

if this is the case, then

the half equivalence point will be lager than normal

will have a higher pH

so

pK = pH ( higher )

meaning that

pKa is highly calculated

so

Ka =10^-pKa = will be LOWE, choose b

C)

We require more data t what Step 2 is... but:

if this is for the base --> no effect

if this is for the acid --> acid gets diluted, therefore accounted for a weaker "acid" expect lower Ka

for both mixtures --> o effect, since both volumes cance each other in Macid*Vacid = Mbase*Vbase

since Vacid = Vbase


Related Solutions

A chemical instrument is calibrated so it has no bias in determining weights of chemical compounds.
Please read the following before answering the next two questions : It turns out that Jack’s data were 10.4, 10.1, 10. 9, 11.0.   A chemical instrument is calibrated so it has no bias in determining weights of chemical compounds. The manufacture claims that it has small random measurement error that follow a normal distribution with mean 0 and standard deviation σ = 0.2 grams. Suppose this claim is true. Two students were assigned to measure the weight of a...
An engineer wants to measure the bias in a pH Meter. She uses the meter to...
An engineer wants to measure the bias in a pH Meter. She uses the meter to measure the pH in 12 neutral substances (pH = 7.0) and obtains the data shown below. 7.01, 7.04, 6.97, 7, 6.99, 6.97, 7.04, 7.04, 7.01, 7.00, 6.99, 7.07 Use the Alpha = 0.05 level of significance to test if the pH meter is calibrated correctly. (a) State the null and alternative hypothesis (b) State your conclusion for your hypothesis test. (c) Give the 95%...
An engineer wants to measure the bias in a pH meter. She uses the meter to...
An engineer wants to measure the bias in a pH meter. She uses the meter to measure the pH in 6 neutral substances and obtains pH readings of 7.08, 7.04, 7.07, 6.99, 7.10, and 7.12. a.   Find the mean and standard deviation for the pH values (show your work). b.   Create a 95% confidence interval for the pH values read by this meter. Use a t critical value of 2.571. c.   Explain in context what your confidence interval tells you...
A process that manufactures glass sheets is supposed to be calibrated so that the mean thickness...
A process that manufactures glass sheets is supposed to be calibrated so that the mean thickness  of the sheets is more than 4 mm. The standard deviation of the sheet thicknesses is known to be well approximated by σ = 0.20 mm. Thicknesses of each sheet in a sample of sheets will be measured, and a test of the hypothesis H_0:μ≤4 versus H_1:μ>4 will be performed. Assume that, in fact, the true mean thickness is 4.04 mm. a) If...
PART I – Prepare a buffer and check the final pH using a pH meter Material...
PART I – Prepare a buffer and check the final pH using a pH meter Material notes: Sodium phosphate, monobasic, NaH2PO4·H2O (Mw = 138) Sodium phosphate, dibasic-pentahydrate, Na2HPO4·7H2O (Mw = 268) pH meter – calibrated using pH 4, 10, 7 reference buffers (done by TA’s) Special safety notes: Wear your personal protective equipment (PPE). Procedure 1. Calculate the amounts of weak acid and conjugate base required to make 50 ml of 0.25 M solutions of each. 2. Prepare 0.25 M...
8. The two methods of determining pH values (pH indicator versus pH meter) should show similar...
8. The two methods of determining pH values (pH indicator versus pH meter) should show similar pH values for those solutions. What was different?
Suppose you performed an exeriment, but forgot to standardize the pH meter. As a result of...
Suppose you performed an exeriment, but forgot to standardize the pH meter. As a result of this oversight, all of the pH readings were incorrectly high. Would your experimentally determined pKa for the solution be incorrectly high, incorrectly low, or unaffected? Explain your answer. Note: pH = pKa + log([conjugate base] / [acid])
Low concentrations of EDTA near the detection limit gave the following dimensionless instrument readings: 175, 104,...
Low concentrations of EDTA near the detection limit gave the following dimensionless instrument readings: 175, 104, 164, 193, 131, 189, 155, 133, 151, and 176. Ten blanks had a mean reading of 50.0. The slope of the calibration curve is 1.75 x 10^-9 M^-1. 1. Estimate the signal detection limit for EDTA. (with correct sig figs) 2. What is the concentration detection limit? (with correct sig figs) 3. What is the lower limit of quantification? (with correct sig figs)
"Low concentrations of dioxin near the detection limit gave the following dimensionless instrument readings: 193.3, 156.5,...
"Low concentrations of dioxin near the detection limit gave the following dimensionless instrument readings: 193.3, 156.5, 150.3, 162.3, 207.3, 148.1, 211.1, 189.7, 122.7, and 193.5. Ten blanks had a mean reading of 47.3. The slope of the calibration curve is 2.55 × 10x9 M–1. Estimate the signal and concentration detection limits and the lower limit of quantitation for dioxin." I know the signal detection limit is yblank + 3s and the concentration limit is 3s/m, but what is ysample to...
1-The following list of readings were taken in sequence during a levelling survey: Readings (m):    1.163,...
1-The following list of readings were taken in sequence during a levelling survey: Readings (m):    1.163, 2.592, 3.616, 0.778, 2.348, 1.838, 2.469, 1.666, 2.893, 1.594    The level instrument was moved after the fifth and eighth readings. Required: (i) Fill in the leveling table and find the RLs of all points. (RL of the first observed point = 105.00m). Use R & F method and check using HCI method. Apply arithmetic check in both cases. *Please solve it by detail.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT