In: Chemistry
An environmental chemist wants to use gas chromatography to measure the herbicide atrazine in water samples. This analysis is originally conducted on a 10-m long and 0.53-mm inner diameter open-tubular column that contains a 1.20-μm thick coating of a stationary phase, which gives a capacity factor of 10.5 for atrazine. The chemist later switches to another column of the same size and type that has a 2.65-μm thick coating of stationary phase. If all other conditions are kept the same, how much will the capacity factor for atrazine change between the old and new column?
ANS Capacity factor of old column = 10.5 for atrazine
as Capacity Factor (K') = (tR - tM)/tM where tR = Retention time which is directly proportional to the coating of a stationary phase and tM is the universal measured of the retention in the chromatography
therefore, tM = (tR- tM)/K' = (1.20 µm - tM )/10.5 = 0.1142 - 0.0952 tM (tR = 1.20µm)
tM (1+0.0952) = 0.1142,
tM = 0.1042 µm
Now, New Capacity factor of column (K') = (tR - tM)/tM
K' = (2.65 µm - 0.1042 µm)/0.1042µm ( New tR = 2.65 µm)
K' = 24.43 (New column capacity factor for atrazine)
Difference Between the capacity factor of old and new column for atrazine will be (24.43 - 10.5 = 13.93) 13.93