Question

In: Chemistry

Low concentrations of Ni-EDTA near the detection limit gave the following counts in a mass spectral...

Low concentrations of Ni-EDTA near the detection limit gave the following counts in a mass spectral measurement: 184, 148, 148, 148, 136, 170., 196, 152, 156, 175. Ten measurements of a blank had a mean of 45 counts. A sample containing 1.00 µM Ni-EDTA gave 1797 counts.

(a) Find the mean.
___ counts

(b) Find the standard deviation.
___ counts

(c) Estimate the detection limit for Ni-EDTA.
___ counts
___ M

Solutions

Expert Solution

(a) The mean (average) of given ten measurements = (184+148 +148 +148 +136 +170 +196 +152 +156+175)/10

Therefore, the mean = 161.3 counts

(b) The standard deviation of given counts (s)for the given measurements is

s = [{(184-161.3)2 + 3*(148-161.3)2 + (136-161.3)2 + (170-161.3)2 + (196-161.3)2 + (152-161.3)2 + (156-161.3)2 + (175-161.3)2}/(10-1)]1/2

s = {(515.29 + 530.67 + 640.09 + 75.69 + 1204.09 + 86.49 + 28.09 + 187.69)/9}1/2.

s = (3268.1/9)1/2

s = (363.12)1/2.

Therefore, the standard deviation = 19.0557 counts

(c) The detection limit for Ni-EDTA (ydl) = yblank + 3*s

i.e. ydl = yblank + 3*s

= 45 + 3*19.0557

Therefore, the detection limit = 102.1671 counts

Conversion of counts to molarity (M)

m = (ysample - yblank) / sample concentration

= (1797 - 45) / 10-6. (since sample concentration = 1 µM, i.e. 10-6 M)

= 1.752 * 109 counts/M

The minimum detectable concentration, c = 3*s/m

c = 3*19.0557/(1.752 * 109), i.e. 3.26296 * 10-8 M


Related Solutions

Low concentrations of EDTA near the detection limit gave the following dimensionless instrument readings: 175, 104,...
Low concentrations of EDTA near the detection limit gave the following dimensionless instrument readings: 175, 104, 164, 193, 131, 189, 155, 133, 151, and 176. Ten blanks had a mean reading of 50.0. The slope of the calibration curve is 1.75 x 10^-9 M^-1. 1. Estimate the signal detection limit for EDTA. (with correct sig figs) 2. What is the concentration detection limit? (with correct sig figs) 3. What is the lower limit of quantification? (with correct sig figs)
"Low concentrations of dioxin near the detection limit gave the following dimensionless instrument readings: 193.3, 156.5,...
"Low concentrations of dioxin near the detection limit gave the following dimensionless instrument readings: 193.3, 156.5, 150.3, 162.3, 207.3, 148.1, 211.1, 189.7, 122.7, and 193.5. Ten blanks had a mean reading of 47.3. The slope of the calibration curve is 2.55 × 10x9 M–1. Estimate the signal and concentration detection limits and the lower limit of quantitation for dioxin." I know the signal detection limit is yblank + 3s and the concentration limit is 3s/m, but what is ysample to...
mass spectral measurements were made on 10 replicate low concentration Ni-EDTA samples: 146.9 , 111.5 ,...
mass spectral measurements were made on 10 replicate low concentration Ni-EDTA samples: 146.9 , 111.5 , 125.7 , 140.7 , 109.7 , 179.1 , 130.7 , 135.3 , 124.7 , and 142.3 counts. Ten measurements of a blank had a mean measurement of 38.3 counts. A sample containing 2.00 μM Ni-EDTA gave a measurement of 1597 counts. Estimate the signal and detection limit for Ni-EDTA.
Calculate the mass concentrations of the following gaseous mixtures enclosed in a room at 2 atmospheres...
Calculate the mass concentrations of the following gaseous mixtures enclosed in a room at 2 atmospheres and 50 oC: (a) Nitrogen (40% by mass), oxygen (40% by mass), and carbon dioxide (20% by mass) in the dry condition. (b) Nitrogen (40% by mass), oxygen (40% by mass), and carbon dioxide (10% by mass), in the presence of 10% H2O by mass. (c) Nitrogen (40%), oxygen (40% by mass), and carbon dioxide (5% by mass), in the presence of 15% H2O...
2. Would the following procedural changes cause the experimentally determined mass percent of NH3 in Ni(NH3)nCl2...
2. Would the following procedural changes cause the experimentally determined mass percent of NH3 in Ni(NH3)nCl2 to be too high, too low, or unchanged. Briefly explain each answer. (a). After dissolving a known mass of Ni(NH3)nCl2, a student directly titrated the NH3 with HCl solution, using a mixed bromocresol green – methyl red indicator. (b). A student added excess standard HCl solution to a known mass of dissolved Ni(NH3)nCl2 and back titrated the excess HCl with standard NaOH solution, using...
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT