Surgical Instrumental: Explain sequence after loading: Start, preheat, prevacuum, water enters, converts to steam, exposure, sterilization...

Surgical Instrumental:

Explain sequence after loading: Start, preheat, prevacuum, water enters, converts to steam, exposure, sterilization

Exhaust, dry, open, unload

Solutions

Expert Solution

Sterilization Cycle Phases

Purge

Steam enters the chamber and displaces ambient air by expelling it through the chamber drain. This displacement can occur by gravity or with use of a vacuum system.

All outlet valves are OPEN during this phase.

Discussion

Air in the chamber acts as a barrier to adequate steam penetration and will negatively affect the sterilization process. A heavily loaded sterilizer will contain many “pockets” where air may become trapped. “Conditioning” aids in the removal of this air, helping to prevent stratification of the functional steam/air mixture.

Conditioning times can be programmed by the operator and can last from 30 seconds to several minutes.

The overall length of the Conditioning phase will depend on the following factors:

Quantity and pressure of the supplied steam.

Density of the goods in the chamber.

Chamber size.

Pre-Vacuum (Vacuum System Required)

Ambient air is efficiently removed from the chamber and the load. This allows steam to completely penetrate the products being sterilized.

This phase is useful for porous instruments or wrapped goods.

Alternative steam and vacuum phases creates pressure/vacuum “pulses”.

Sterilization (Dwell)

Interior temperature and pressure are maintained at setpoint for a time specified by the controller or manual protocol.

Drying (Cool Down)

The interior load returns to ambient temperature and pressure in any of three typical methods.

Gravity (Fast Exhaust)

This method allows the steam and condensate to exit through the chamber drain with no mechanical assistance. As the effluent discharges, the pressure and temperature in the chamber decrease. When the chamber pressure is near zero, the door can be opened and the goods may be removed. This method is simple, but not the fastest.

Gravity (Slow Exhaust)

For liquids. The cool down is gradual to permit dissolved gases in the liquid media to reach equilibrium; this avoids boilover until ambient temperature and pressure are restored.

Liquid, air-over-pressure. The cool down is gradual as pressurized air is injected to replace steam, thereby lowering the interior temperature under pressure and preventing smaller media samples from evaporating. Once temperature is reduced, air pressure is gradually equalized and ambient temperature and pressure are achieved.

Vacuum Drying (Post Vacuum)

A vacuum system pulls the steam and condensate through the drain port. The longer the vacuum system runs during the dry phase, the cooler and dryer the goods will be when removed from the chamber.

The Vacuum exhaust is programmed to run for a settable amount of time.

Vacuum exhaust should not be used on sterilized liquid because a quick reduction in pressure will cause the liquid to boil over.

Sterilization Cycle Parameters

Many first-time sterilizer users want to know how to select the best cycle parameters for their applications. The answer depends on at least four factors:

Size of load

Type of load (waste, media, glassware, or other)

Time and throughput constraints

Steam source (house steam, integral steam generation or other)

PLEASE DO LIKE??


Related Solutions

Surgical Instrumental: Explain displays for wrapped items, uncovered items, peel pouches, dental drills.
Surgical Instrumental: Explain displays for wrapped items, uncovered items, peel pouches, dental drills.
A steam thermoelectric plant operates a Rankine cycle with overheating. Water enters the pump as saturated...
A steam thermoelectric plant operates a Rankine cycle with overheating. Water enters the pump as saturated liquid , 7.7 kg/s. The efficiency of the pump is 0.9 The water enters the high pression turbine at 12 500 kPa and 550°C, after that steam overheats at a constant pressure of 2000 kPa until 450°C before it expands on the low pressure turbine. The efficiency of both turbines is 0.85. The vapor quality when it enters to the condenser is 0.95 Determine...
A steam thermoelectric plant operates a Rankine cycle with overheating. Water enters the pump as saturated...
A steam thermoelectric plant operates a Rankine cycle with overheating. Water enters the pump as saturated liquid , 7.7 kg/s. The efficiency of the pump is 0.9 The water enters the high pression turbine at 12 500 kPa and 550°C, after that steam overheats at a constant pressure of 2000 kPa until 450°C before it expands on the low pressure turbine. The efficiency of both turbines is 0.85. The vapor quality when it enters to the condenser is 0.95 Determine...
A reheat Rankine cycle operates with water as the working fluid. Steam enters the first turbine...
A reheat Rankine cycle operates with water as the working fluid. Steam enters the first turbine at 8 MPa and 450◦C and exits at 0.8 MPa. It is then reheated at 400◦C before entering the second turbine, where it exits at 10 kPa. If the amount of work into the pump is 8.04 kJ/kg and the net work per cycle produced is 1410.5 kJ/kg, determine the thermal efficiency. Assume no pressure losses in the condenser or the boiler.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT