Question

In: Chemistry

Drug money extract in the 5.0 mL methanol containing cocaine gave a signal of 4.27 mV...

Drug money extract in the 5.0 mL methanol containing cocaine gave a signal of 4.27 mV in a GC-FID experiment. Then 500 L of 20.0 ppm cocaine were added to the 5.0 mL methanolic extract. This spiked solution sample gave a signal of 7.98 mV. Find the concentration of cocaine in the original extract, in ppm, in mol/L, in g.

Solutions

Expert Solution

molar mass of cocaine = 303.353 g/mol

molarity of 20 ppm solution of cocaine = 20 mg/L = 20/303.353 = 0.066 mmol/L

moles of cocaine added = molarity x volume

                                       = 0.066 mmol/L x 0.500 L

                                       = 0.033 mmol

Molarity of cocaine in 505 mL (0.505 L) solution = moles/L

                                       = 0.033/0.505

                                       = 0.065 mmol/L

We have 0.065 mmol/L solution gave a signal of 7.98 mV in GC-FID

So, 4.27 mV would be for a cocaine concentration = 4.27 x 0.065/7.98

                                        = 0.035 mmol/L in 5 mL of sample solution

Concentration in mol/L = 3.5 x 10^-5 mol/L

Concentration in ppm = 0.035 x 303.353

                                        = 10.62 mg/L

                                        = 10.62 ppm

Concentration in g = molarity x volume x 303.353

                                         = 0.035 mmol/L x 0.005 L x 303.353 g/mol x 1/1000

                                         = 5.31 x 10^-5 g

                                        = 0.01062 g


Related Solutions

Analyte in an unknown gave a signal of 25.0 mV. When 1.00 mL of 0.065 M...
Analyte in an unknown gave a signal of 25.0 mV. When 1.00 mL of 0.065 M standard addition reagent was added to 10.00 mL of the unknown sample, the signal increased to 35.0 mV. What is the analyte concentration of the original unknown?
10.00 ml of a solution of an analyte gives a signal of 30 mV in an...
10.00 ml of a solution of an analyte gives a signal of 30 mV in an instrumental analysis. after the addition of 2e-5 mole standard, the signal rises to 75 mV. what is the concentration of the analyte in the original solution? assuming the total volume doesn't change.
A 25.0 mL sample containing Ni2+ was analyzed and gave an instrument siganl of 23.6 units...
A 25.0 mL sample containing Ni2+ was analyzed and gave an instrument siganl of 23.6 units (corrected for a blank). When exactly 0.500 mL of 0.0287 Ni(NO3)2 was added to the solution, the signal increased to 37.9 units. Calculate the molar concentration of Ni2+ assuming that the signal was directly proportional to the analyte concentration. I was told this was a single addition standard addition problem, but I'm not sure how to work it. Thanks!
Calculate the initcial I^- concentration in a solution containing 3.0 mL of 0.20 M KI, 5.0...
Calculate the initcial I^- concentration in a solution containing 3.0 mL of 0.20 M KI, 5.0 mL of 0.15 M (NH4)2S2O8, 5.0 mL of 0.25 M (NH4)SO4, 6.0 mL of 0.10 M KNO3, and 1.0 mL of 0.012 M Na2S2O3.
A hospital pharmacist prepares an intravenous solution containing 300 mg of DRUG X in 250 mL...
A hospital pharmacist prepares an intravenous solution containing 300 mg of DRUG X in 250 mL normal saline. The infusion rate of DRUG X is 15 mcg/kg/min, and this solution is to be administered to a patient weighing 132 pounds. What would be the appropriate infusion rate of the solution for this patient in mL/hr? The rate is _______ mL/hr.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT