Question

In: Chemistry

An open tubular column with a diameter of 203 μm and a stationary phase thickness on...

An open tubular column with a diameter of 203 μm and a stationary phase thickness on the inner wall of 0.75 μm passes unretained solute through in 53 s. A particular solute has a retention time of 435 s.

a.) What is the partition coefficient for this solute?

K=_____

b.) What fraction of time does this solute spend in the stationary phase?

Ts/Ttotal=_____

Solutions

Expert Solution

solution:

b)fraction of time spent in stationary phase = time spent in stationary phase/total time spent

fraction of time spent in stationary phase = (435 – 53 s)/435 s = 0.878 = 87.8%

a)k = (435 – 53 s)/53 s = 7.2

K = Cs/Cm = k(Vm/Vs)

Vm = Vcylinder = π(d2/4)L = π[(0.203 mm)2/4)L = 0.0324 mm2L

Vs = Vtube = Vouter cylinder – Vinner cylinder = π[(0.203 mm)2/4]L – π{[(0.207 mm – 2(0.00075 mm)]2/4}L

Note: for a thin tube, Vtube = πtdL, where t = film thickness. This same equation also can be calculated from the first equation (Vs = Vtube = Vouter cylinder – Vinner cylinder ) since

Vtube = π[(0.203 mm)2/4]L – π[(0.203 mm)2/4]L + π(0.203 mm)(0.00075 mm)L –

π(0.0075 mm)2L. The first two terms cancel out and since 0.203 mm >> 0.00075 mm, the last term is inconsequential.

Vs = Vtube = π(0.203 mm)(0.00075 mm)L = 0.0005 mm2L

K = k(Vm/Vs) = 7.2(0.0324 mm2L/0.0005 mm2L) = 467 (note how L cancels out so the column length is not needed).


Related Solutions

An open tubular column with a diameter of 193 um and a staionary phase thickness on...
An open tubular column with a diameter of 193 um and a staionary phase thickness on the inner wall of 0.52 um passes unretained solute through in 61 s. A particular solute has a retention time of 453 s. a) What is the partition coeffecient for this solute? b) What fraction of time does this solute spend in the stationary phase?
What are open tubular columns and give some differences between open tubular column and packed column
What are open tubular columns and give some differences between open tubular column and packed column
compare column, thin-layer, and gas chromatography in terms of mobile phase and stationary phase
compare column, thin-layer, and gas chromatography in terms of mobile phase and stationary phase
With respect to liquid column chromatography in which the stationary phase is silica, propose a suitable...
With respect to liquid column chromatography in which the stationary phase is silica, propose a suitable strategy for separating two components of a binary mixture that fail to move when hexane is the mobile phase and move the same distance when methanol is the mobile phase. Explain your reasoning.
Consider an optical fiber that has a 200-μm core diameter and a 240-μm cladding diameter. (a)...
Consider an optical fiber that has a 200-μm core diameter and a 240-μm cladding diameter. (a) Show that the size of the active area in a bundle with two rings is 1.13 mm2. (b) Show that the ratio of the active area to the total cross sectional area of the bundle is 53 %.
Determine the time required to obtain a Cu deposit of thickness 1 μm when electrodeposition is...
Determine the time required to obtain a Cu deposit of thickness 1 μm when electrodeposition is done at 4 A and 6 A, respectively. The surface area of the substrate is 314 cm2.
Which phase carries the analyte through the Chromatographic separation? Standard Phase, Detection Phase ,Injection Phase, Stationary...
Which phase carries the analyte through the Chromatographic separation? Standard Phase, Detection Phase ,Injection Phase, Stationary Phase, Mobile Phase Which phase slows the analyte down in the Chromatographic separation? Standard Phase, Detection Phase ,Injection Phase, Stationary Phase , Mobile Phase If you increase the flow rate of the mobile phase, will the analyte come out ________? Slower, In two Peaks, Faster, Unchanged In this separation the stationary phase is polar and the mobile phase is non-polar. Youre separating two molecules,...
1) The kidneys are placed ________________ to the vertebral column 2) •Spinal Cavity The tubular space...
1) The kidneys are placed ________________ to the vertebral column 2) •Spinal Cavity The tubular space formed by the vertical stacking of the individual ________________.
What are the surface functional groups on the stationary phase in normal phase LC and reverse...
What are the surface functional groups on the stationary phase in normal phase LC and reverse phase LC. Write out the eluotropic series for both normal phase LC and reverse phase LC using the common solvents we discussed in class (acetonitrile, MeOH, diethylether, H2O, pentane, toluene).  Why acetonitrile is a stronger eluting solvent than toluene in normal phase LC?
During which phase of growth would a primary metabolite be produced? lag phase exponential phase stationary...
During which phase of growth would a primary metabolite be produced? lag phase exponential phase stationary phase death phase all are correct
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT