In: Chemistry
In the lanthanum fluoride
electrode, the sensing element is a crystal of lanthanum fluoride
LaF3, doped with europium fluoride EuF2 to create lattice
vacancies. Such a crystal is an ionic conductor by virtue of the
mobility of fluoride ions which jump between lattice vacancies. An
electrochemical cell may be constructed using such a crystal as a
membrane separating two fluoride solutions. This cell acts as a
concentration cell with transference where the fluoride transport
number is 1. As transference of charge through the crystal is
almost exclusively due to fluoride, the electrode is highly
specific to fluoride. The only ion which significantly interferes
is hydroxide (OH−). Generally such "alkaline error" can be avoided
by buffering the sample to a pH below 7.
Cell diagram[edit]
The cell diagram of a typical experimental arrangement is:
Ag,AgCl|NaF|LaF3|fluoride analyte solution
where:
Hg|1/2Hg2Cl2 | KCl(sat) is an external reference electrode
NaF,NaCl | AgCl,Ag | is an internal reference inside the fluoride
selective electrode
Reference electrode[edit]
Some commercially available reference electrodes have an internal
junction which minimizes the liquid junction potential between the
sample solution and the electrolyte in the reference electrode
(KCl). The internal electrolyte is at fixed composition and the
electrode response is given by the Nernst equation:
E = E0 − RT/F ln aF−,
where:
E is the measured cell potential,
E0 is the standard cell potential,
R is the ideal gas constant,
T is the temperature in kelvins,
F is the Faraday constant (9.6485309×104 C/mol).
aF− is the activity of the fluoride ion.