Question

In: Other

On a chromatographic column one meter in length, operating with an efficiency of 10,000 plates, the...

  1. On a chromatographic column one meter in length, operating with an efficiency of 10,000 plates, the retention time of ?-chlorestne and ?-chlorestne are 4025 and 4100 seconds respectively. If these two compounds are to be separated with baseline resolution, can the desired separation be accomplished using this one meter column? The dead time of the column is 480 seconds.

Solutions

Expert Solution

Given:

N =Number of theorotical plates = 10000 plates

trA = retention time of A-Chlorestne = 4025 seconds

trB = reention time of B-chlorestne = 4100 seconds

to = dead time of the column = 480 seconds

Here in this problem we have to check foe desired separation can be achieved by using one meter column.

So first we find the peak width at half height by formula given as,

  

for A-chlorestne = 161 seconds and = 164 seconds

The degree of separation between two compounds can be calculated by formula as,

On substituting the values we get,

In general we want according to standard results,

We have one more standard formula as,

Here we can say from the formula that the   

Let R1= 0.461 , R2 = 1.5 and  N1 = 10000,   

Solving the above equation, we get

Height equivalent to theorotical plate is given by

So initially using 10000 plates and length one meter we get the H= 1 * 10-4 m

Now, we know that   .....Ans

So desired separation cannot be achieved by one meter column. For this we will need column of height to achieve desired separation.


Related Solutions

A chromatographic analysis is performed with a sample containing toluene and benzene. The column length was...
A chromatographic analysis is performed with a sample containing toluene and benzene. The column length was 10 m and the flow rate 30 mL/min. Solute               tR                        W1/2 air                      1.5 benzene (A)     7.45                   1.05 toluene (B)       10.60                 1.45 What is the resolution for the two peaks? 1.26 0.25 0.04 2.52
What will be the length (L in meter) of a rectangular footing to support rectangular column...
What will be the length (L in meter) of a rectangular footing to support rectangular column 20x70 cm and carries a dead load of 625 kN and live load of 426 kN. The net allowable bearing pressure is 155 kPa. The design should be according to (ACI) code with the following parameters: fc’ = 28 MPa, fy = 420 MPa.
Which of the following will increase the number of theoretical plates in a chromatographic separation? Select...
Which of the following will increase the number of theoretical plates in a chromatographic separation? Select all that are appropriate. decrease column packing particle size increase column mobile phase flow rate increase column length change mobile phase to one with a greater analyte diffusion coefficent
You are asked to run a chromatographic column for separation of a mixture of compounds: CH3CH2OCH3,...
You are asked to run a chromatographic column for separation of a mixture of compounds: CH3CH2OCH3, CH3(CH2)3OH, CH3CO2H, CH3(CH2)8CH3. Two solvents have been provided i.e. acetone and hexane. Which eluting solvent would you use first and predict the order of elution for the compounds.
A chromatographic column in which oligo-dT is linked to an inert substance is useful in separating...
A chromatographic column in which oligo-dT is linked to an inert substance is useful in separating eukaryotic mRNA from other RNA molecules. On what principle does this column operate?
The following conditions were established for a chromatographic separation of multiple hydrocarbons on a packed column...
The following conditions were established for a chromatographic separation of multiple hydrocarbons on a packed column with 1.1 g of liquid stationary phase: inlet pressure 826 torr, room pressure 756 torr, vapor pressure of water at room temperature 17.5 torr; column temperature 110ºC, room temperature 20°C; measured outlet flow rate 20.0 mL/min; retention time for nonretained species 25.3 s, retention time of the first eluting species 1.56 min. Based on these conditions, the pressure drop correction factor j is _______,...
Please outline only the main steps of an efficient column chromatographic separation of a non-polar compound...
Please outline only the main steps of an efficient column chromatographic separation of a non-polar compound A from a medium polar compound B and a non-polar chemical (C). C is slightly less polar than A.
The following data apply to apply to a column for liquid chromatography:                    Length of column:...
The following data apply to apply to a column for liquid chromatography:                    Length of column: 25.7 cm                    Flow rate:            0.313 mL/min                    VM:                      1.37 mL                    VS:                       0.164 mL A chromatogram of species A and B provided the following data: Retention Time, min Width of Peak Base (W), min Non-retained 3.1 - A 13.3 1.07 B 14.1 1.16 a)    the expected resolution, if the column length was doubled b)    the expected capacity factors,...
What will be the width (B in meter) of a square footing to support rectangular column...
What will be the width (B in meter) of a square footing to support rectangular column 40x40 cm and carries a dead load of 614 kN and live load of 477 kN. The net allowable bearing pressure is 180 kPa. The design should be according to (ACI) code with the following parameters: fc’ = 28 MPa, fy = 420 MPa.
1- Brief discussion of column chromatographic separation (solvent polarity, elution of compounds etc.) 2- Relationship between...
1- Brief discussion of column chromatographic separation (solvent polarity, elution of compounds etc.) 2- Relationship between polarity and thin layer chromatography​.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT