The following data was collected from a TLC plate for 4 different compounds. The distance's traveled by each is shown below:

Spot A: 8.2 cm

Spot B: 0.7 cm

Spot C: 6.5 cm

Spot D: 1.3 cm

The spotting line was placed 1.5 cm above the base and the solvent traveled 10.5 cm from the bottom of the plate. Calculate the Rf value for each compound.

My answers to this were

A: 0.78 B: 0.067 C: 0.62 D: 0.12

I got this question wrong, can someone explain why?

Solutions

Expert Solution

TLC stands for Thin Layer Chromatography; it is an analytical technique in which we exploit each sample's affinity toward different polar/nonpolar solvents.

We use a stationary phase: a phase which does NOT moves, as the name implies. It is typically the "thin layer" or the "plate". It is commonly to use silica gel plates, which are polar, but any species could be used, such as paper.

We also use a mobile phase: as the name implies, this phase will "move" and will carry along the path all the components from the solvent. Many times, you will be able to split in several components the "mixture" of the sample. Therefore, they can be seen in the stationary phase (plate). Typically, we allow the mobile phase to go up to the "Rfinal" mark, or final reference. This is the distance travelled by the solvent (mobile phase). It starts at solvent level, it is R0, therefore.

Rf = distance traveled by sample / distance traveled by solvent

Then, "Rf" value allows us to understand the affinities of anaytes towards the solvent.

As Rf increases, this implies higher affinity toward the mobile phase ( since it is carried all along the mobile phase)

If Rf is low, this implies a low affinity toward the mobile phase; therefore higher affinity toward the stationary phase, since it sticks to the plate rather than travel upwards.

now...

Solvent Front = (10.5-1.5) = 9 cm

so

Rf-A = 8.2/9 = 0.911

Rf-B = 0.7/9 = 0.077

Rf-C = 6.5/9 = 0.722

Rf-D = 1.3/9 = 0.144

your probable error --> you are using 10.5 as a basis, not 10.5-1.5 M which is the actual solvent distance


Related Solutions

1. Which amino acid traveled the furthest up the TLC plate and which amino acid traveled...
1. Which amino acid traveled the furthest up the TLC plate and which amino acid traveled the least? 2. Why must use a pencil instead of pen to write on TLC plate? 3. What is the function of ninhydrin spray? 4. What are the amino acid present in the unknown solution?
In TLC, the polarity of the different compounds to be analyzed is highly critical. Which one...
In TLC, the polarity of the different compounds to be analyzed is highly critical. Which one of the following statements is true? (Check all that apply) 1 The less polar components of the mixture to be analyzed will be less attracted to the stationary phase and therefore move further up the plate. 2. Any solvent mixture can be used to separate any compound mixture with TLC. 3. The more polar components of the unknown mixture move faster up the plate...
A mixture of 4-methylphenol and 4-methylcyclohexanol was run on a TLC plate using 3:1 hexane:ethyl acetate...
A mixture of 4-methylphenol and 4-methylcyclohexanol was run on a TLC plate using 3:1 hexane:ethyl acetate as the eluent. Only one spot was seen using a UV lamp. Does this mean that both spots have the same Rf or is there another explanation? How might you solve this problem?
Explain for each of the following methods how they visualize compounds on TLC plates. Which types...
Explain for each of the following methods how they visualize compounds on TLC plates. Which types of compounds are not detectable by each method? a) Fluorescence quenching b) iodine treatment
Briefly explain the possible effect of the following experimental errors. TLC plate procedure with drug concentration....
Briefly explain the possible effect of the following experimental errors. TLC plate procedure with drug concentration. 1. You used a 1:200 ratio of ethyl acetate/acetic acid rather than 200:1: (Note: ethyl acetate is significantly less polar than acetic acid) 
2. You allowed the TLC plate to develop too long (past the point where the solvent reaches the top of the plate). 3. You used a plastic pipet rather than a thin glass spotter. 4. You labeled your plate using a...
1) The following data were collected from a repeated-measures study investigating the effects of 4 treatment...
1) The following data were collected from a repeated-measures study investigating the effects of 4 treatment conditions on test performance. Determine if there are any significant differences among the four treatments. State the null hypothesis. If you determine a significant treatment effect, use Tukey’s HSD test (overall α = .05) to determine which treatments differ from which other treatments. Also, compute the percentage of variance explained by the treatment effect (η2). Conclude with an appropriate summary describing what you found....
1. The following data shows memory scores collected from adults of different ages. Age (X) Memory...
1. The following data shows memory scores collected from adults of different ages. Age (X) Memory Score (Y) 25 10 32           10 39 9 48 9 56 7 Use the data to find the regression equation for predicting memory scores from age Group of answer choices a. Ŷ = 4.33X + 0.11 b. Ŷ = -0.11X + 4.33 c. Ŷ = -0.11X + 13.26 d. Ŷ = -0.09X + 5.4 e. Ŷ = -0.09X + 12.6 1a Use the regression...
What technique can be used to monitor the compounds that are collected from the column?
What technique can be used to monitor the compounds that are collected from the column?
The following data were collected in an experiment designed to investigate the impact of different positions...
The following data were collected in an experiment designed to investigate the impact of different positions of the mother during ultrasound on fetal heart rate. Fetal heart rate is measured by ultrasound in beats per minute. The study includes 20 women who are assigned to one position and have the fetal heart rate measured in that position. Each woman is between 28-32 weeks gestation. The data are shown below. Back Side Sitting Standing 140 141 144 147 144 143 145...
The following data were collected in an experiment designed to investigate the impact of different positions...
The following data were collected in an experiment designed to investigate the impact of different positions of the mother during ultrasound on fetal heart rate. Fetal heart rate is measured by ultrasound in beats per minute. The study includes 20 women who are assigned to one position and have the fetal heart rate measured in that position. Each woman is between 28-32 weeks gestation. The data are shown below. Back Side Sitting Standing 140 141 144 147 144 143 145...
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT