Parties to the conflict / type of medical service: Medical physicist, radiotherapist, operator (nurse-apparatus), topometrist / external radiation therapy on a gamma-therapy device
I. According to the treatment of the patient on the gamma-therapeutic apparatus according to the Regulations of the department, the patient undergoes topometric preparation.
Mandatory participants in patient topometric
preparation:
a doctor - a radiotherapist in the radiation and topometric
preparation maps indicates the location and apparatus for
treatment, determines the set of additional fixation devices to the
base set according to the approved irradiation method, selects a
convenient position for the patient taking into account the
localization and condition of the patient, fills in the patient
registry, topometric preparation protocol and transmits information
to the topometrist (radiologist);
the topometrist doctor (radiologist)will coordinate with the
radiotherapista set of additional fixing devices to the base set
and make them as needed (thermoplastic masks), and also selects the
patient’s convenient position taking into account the locations and
the set of fixing devices used, then completes the topometric
protocol preparation for radiation therapy and checks the number of
images from a computer tomograph (CT images);
the medical physicistselects the desired position of the patient,
taking into account the technical characteristics of the
therapeutic unit, adjusted for the geometry of the fields at which
the treatment will be carried out.
II. After topometric preparation, digital X-ray images from a
Toshiba Aquilion computer tomograph are transmitted to the Monaco
Sim contouring station. Radiotherapistchecks the number of CT
images (not more than 300 according to the requirements of the
contouring system), performs contouring of the target (GTV, CTV)
and the planned exposure volume (PTV), organs of risk (OAR) in
accordance with the literature.
contouring for conventional radiation therapy (CRT) to prepare 1
working day;
Note: In the case of multi-stage treatment, PTVs
are allocated for each stage of treatment with the corresponding
value of the total dose for this stage of planning.
III. Next, the radiotherapist fills out a prescription protocol
with a description of the course of radiation therapy: name of the
patient, diagnosis, patient ID, area of exposure, irradiation
technique (Conventional RT), therapeutic unit, total dose per RT
course, fractionation mode, and also necessary conditions for
target coverage in accordance with the objectives of RT:
Conventional RT: At least 85% -90% of the prescribed dose should
be irradiated with at least 90% of the planned exposure. No more
than 107% of the prescribed dose can be irradiated with no more
than 5% of the planned exposure. Radiotherapist prescribes
restrictions on risk organs in accordance with the literature and
fills out a prescription protocol.
Note:All calculations of the isoeffective dose for
the planned volume of exposure and risk organs in the case of using
a fractionation regimen other than the standard regime, as well as
interruption of the radiation therapy course, should be carried out
according to the linear-quadratic radiobiological model
according to the literature.
IV. A radiotherapistconfirms in writing with the head. or senior
physician - radiotherapist, contours of anatomical structures and
prescription protocol.
V. Radiotherapistcompletes treatment in the MOSAIQ administrative
system:
Select a patient and open the Diagnoses and Invasive Procedures
tab;
Establish a patient diagnosis;
Create a prescription for radiation therapy (indicate the name of
the area; prescribed dose, number of fractions and fractional
mode);
For each stage of treatment, a prescription is created.
VI. Radiotherapistverifies the fulfillment of the necessary points
according to the Checklist on the preparation of a patient for
remote radiation therapy approved by the radiotherapy department on
the Theratron Equinox 100 gamma apparatus.
VII. Radiotherapistsends contoured CT images to the planning
station (XIO) and makes an entry in the journal “Application for
planning of radiotherapy”.
VIII. Medical physicist checks the fulfillment of the necessary
points according to the Checklist for preparing the patient for
remote radiation therapy for the Theratron Equinox 100 gamma-ray
apparatus approved in the radiotherapy department, the presence of
contoured CT images, the approved prescription and entry in the
journal “Application for planning of radiotherapy”.
IX. Medical physicistcreates a dosimetric treatment plan. If
necessary and the complexity of the location of the localization
creates several dosimetric treatment plans for the possible
selection of the most optimal in accordance with the
recommendations. Options for dosimetric plans are discussed
between
medical physicists of the Department of Medical Physics (DMP). A
plan is created for the entire course of radiation therapy.
dosimetric planning for conventional radiation therapy
(Conventional RT) to prepare 1 working day;
Note: All calculations of the isoeffective dose
for the planned volume of exposure for the area of nonirradiation
(“cold spots”) and the area of overexposure (“hot spots”) of the
isodose distribution, as well as for risk organs with an area of
high isodose lines, should be carried out according to the linear
quadratic radiobiological model according to the literature.
X. Medical physicist and radiotherapist approve the optimal
exposure plan according to the following criteria:
FOR TARGET:
According to the histogram of the distribution of dose - volume
DVH:
at least 85% -90% of the prescribed dose should be irradiated
with at least 90% of the planned exposure. No more than 107% of the
prescribed dose can be irradiated with no more than 5% of the
planned exposure;
According to the parameters of the statistical dose
distribution;
By visual inspection of the scores distribution of percentage
isodose lines (107%, 100%, 95%, 90%, 85%, 80% 50%, 20%) relative to
the prescribed total dose for the entire course of radiation
therapy.
FOR ORGAN AT RISK (OAR):
the maximum dose in the OAR should not exceed x Gy not more than
v% of the volume of ROI can receive x Gy (Dv%≤ x);
isoeffective doses for "hot spots" are calculated according to
the linear-quadratic radiobiological model according to the
literature;
according to the international QUANTEC and RTOG protocols, if a
risk organ with a third level of priority is not passed, the
radiotherapist decides to accept or reject this treatment
plan;
for any deviations according to the statistical data of the dose
distribution from the prescribed, the doctor also makes a decision
on the choice or deviation of this plan and fixes his decision in
the radiation map.
XI. The final version of the patient’s dosimetric plan for
irradiation is approved at the general meeting of the radiotherapy
department in accordance with the Regulations of the radiotherapy
department and signed by the head of Medical Physics Department
(MPD) and head of the radiotherapy department.
XII. Medical physicistenters in the patient’s radiation card the
following necessary documentation:
exposure protocol (geometric and dosimetric parameters of the
exposure plan), which is signed by the second medical
physicist;
histogram dose volume (DVH) for the required set of anatomical
structures;
statistical dose distribution for the entire set of anatomical
structures;
shift of the radiation field relative to reference marks to
determine the isocenter of the treatment field on the patient and
the displacement of the therapeutic table with the patient relative
to the isocenter of the installation.
XIII. Medical physicistenters the following necessary information
and documentation to the MOSAIQ system:
exports information on the created treatment plan (parameters of
each radiation field, statistical dose distribution for each
structure and dose distribution depending on the volume of the
structure in the form of a DVH histogram) as a file with the
extension.pdf.
XIV. Medical physicistchecks the fulfillment of the necessary
points according to the Checklist on the preparation of the patient
for remote radiation therapy approved by the radiotherapy
department on the Theratron Equinox 100 gamma apparatus.
XV. Topometrist on a computed tomograph, the patient is displaced
along the laser system relative to the reference marks to the
isocenter position of the treatment plan.
XVI. Topometristenters the patient into the schedule of operation
of this device, confirms in the MOSAIQ system the treatment courses
and prescription of the patient, as well as the area of
exposure.
XVII. Radiotherapisttransmits documentation (radiation card) for
the patient to the nurse operator.
XVIII. Operatorchecks the fulfillment of the necessary points
according to the Checklist for the preparation of a patient for
remote radiation therapy approved by the radiotherapy department
for the Theratron Equinox 100 gamma apparatus and the availability
of all necessary documentation in the radiation card.
XIX. When the patient is first laid on the therapeutic unit, the
following actions are performed with the participation of the
following responsible persons:
1) Radiotherapist- together with the operator, they prepare and lay
the patient on the treatment table of the therapeutic unit
according to the established position marks of the isocenter of the
dosimetric plan, which were installed on a computer
tomograph;
2) Topometrist- ascertains the correct position of the patient and
the use of an appropriate set of fixing devices according to the
protocol with topometric preparation;
3) Medical physicist - makes sure that the created dosimetric
treatment plan (patient position and radiation field) is correctly
reproduced on the patient;
4) The operator conducts irradiation;
5) If the plan contains several isocenters and the stage of
treatment is changing, then it is necessary to carry out the
procedure in accordance with paragraphs 1)-4).
Note:In case of technical malfunctions of the
therapeutic unit, as well as dosimetric changes in the radiation
plan, the engineer / physical engineer / medical physicist of
MPD.
XX. In subsequent sessions of radiation therapy, the patient is
laid on the treatment table and treatment at
the appropriate installation is carried out by the operator.
Note: n the event of a change in the patient's
laying parameters at the therapeutic unit and the general
condition of the patient, a radiotherapistis called.
XXI. In the event of a change in the patient’s position on the
therapeutic unit during the course of RT, it is necessary to carry
out the procedure in accordance with I-ХIX.
1. What is the assessment of time savings (device
resources) in accordance with this regulation?
2. How much does accuracy increase in the delivery of the
prescribed dose, increasing the effectiveness of RT?
3. How will this regulation affect the regulations on the Quality
Assurance of the apparatus and topometry?
4. How will this regulation affect the qualifications and
compactness of each employee?
Brachytherapy and Orthovoltage
Orthovoltage device has mosfet tubes and beryllium windows which makes it partially expensive.
Brachytherapy device also come in varients based on the region of interest. There are seperate devices for skin, breast tissue, rectum and GI tumors. There are also much advanced varients in intrabeam and photoelectric therapy and they are effectively costly. Zeiss lens for intrabeam is much costlier. Voltage based varients also exist and they are based on the capacity and usage characteristics
2. Accuracy in delivery: Beam align verification in conventional radiotherapy are visualized through port films and electronic portal imaging. In Orthovoltage RT there is non uniform dose disribution and maximum dose is applied on the skin. Dose distribution is apparantly non-present in Conventional RT.
Delivery verificationa nd modification is adept and is done by pulse by pulse basis and hence there is much more accuracy in Brachyherapy RT devices.
3. Effect on the regulations on quality assurance:
Feasibility sub studies and toxic assessment is performed on more number of random patients in coventioanl RT. Example: A thoracic SABR safety sub study had minimum of 50 patients and unacceptable deviation is much higher
For Brachytherapy and Orthovoltage there is comparatively more number of acceptable variation and less deviation per protocol
4. Impact of the tech specifications on each employee:
For conventional radiography a radiophysist and a physician are compatible to work with the machine. Brachytherapy and Orthovoltage devices need a tomographist and a radiophysist for effective treatment, Adjustability of the volatages and the device handling should be professional and any error through deviation proves costly.