Describe the 3 types of membranes commonly used to separate the test solution from the internal...

Describe the 3 types of membranes commonly used to separate the test solution from the internal reference electrode.

Solutions

Expert Solution

Different membrane materials have proved to give optimal responses for certain species. For example, a glass membrane is unsurpassed for measuring H+ activity, pH. This ISE may be referred to as a glass or pH electrode.

Liquid membrane electrodes have non-glass, solid-state crystals or pellets as the membrane component of the electrode. This approach has proved effective for numerous cations and anions. The most successful example is the excellent electrode for fluoride ion, which is based on a crystal of LaF3 doped with Eu(II) to create crystal defects to improve conductivity.

Gas-sensing electrodes or combination electrodes that respond to the concentration of gases dissolved in aqueous solution. The electrodes consist of an ion-selective electrode, usually pH, in contact with a thin layer of solution that is held in place with a membrane permeable to the desired gas such as NH3 or CO2. When the gas passes through the membrane, the change of pH in the thin layer of solution is sensed by the glass membrane pH electrode.

Other combination electrodes consist of an enzyme immobilized on an ISE. The ISE is chosen to respond to a product of the enzyme-substrate reaction and the selectivity is provided by the enzyme.

You can read the text below for more information about the membranes and electrodes. [Source: NMSU]

a) The Silver / Silver Chloride Single Junction Reference Electrode.

The most common and simplest reference system is the silver / silver chloride single junction reference electrode. This generally consists of a cylindrical glass tube containing a 4 Molar solution of KCl saturated with AgCl. The lower end is sealed with a porous ceramic frit which allows the slow passage of the internal filling solution and forms the liquid junction with the external test solution. Dipping into the filling solution is a silver wire coated with a layer of silver chloride (it is chloridised) which is joined to a low-noise cable which connects to the measuring system.

In electrochemical terms, the half-cell can be represented by:

Ag / AgCl (Satd), KCL (Satd)

and the electrode reaction is:

AgCl (s) + e- = Ag (s) + Cl-

The electrode potential for this half-cell is + 0.2046 V relative to the Standard Hydrogen Electrode at 25°C

b) Double Junction Reference Electrodes.

One problem with reference electrodes is that, in order to ensure a stable voltage, it is necessary to maintain a steady flow of electrolyte through the porous frit. Thus there is a gradual contamination of the test solution with electrolyte ions. This can cause problems when trying to measure low levels of K, Cl, or Ag, or when using other ISEs with which these elements may cause interference problems. In order to overcome this difficulty the double junction reference electrode was developed. In this case the silver / silver chloride cell described above forms the inner element and this is inserted into an outer tube containing a different electrolyte which is then in contact with the outer test solution through a second porous frit. The outer filling solution is said to form a "salt bridge" between the inner reference system and the test solution and is chosen so that it does not contaminate the test solution with any ions which would effect the analysis.

Commonly used outer filling solutions are:

potassium nitrate - for Br, Cd, Cl, Cu, CN, I, Pb, Hg, Ag, S, SCN.

sodium chloride - for K,

ammonium sulphate - for N03,

magnesium sulphate - for NH4,

Note that double junction reference electrodes are named after their outer filling solutions.

One disadvantage with double junction reference electrodes is that they introduce an extra interface between two electrolytes and thus give the opportunity for an extra liquid junction potential to develop.

c) Liquid Junction Potentials.

It must be noted that the standard voltage given by a reference electrode is only correct if there is no additional voltage supplied by a liquid junction potential formed at the porous plug between the filling solution and the external test solution. Liquid junction potentials can appear whenever two dissimilar electrolytes come into contact. At this junction, a potential difference will develop as a result of the tendency of the smaller and faster ions to move across the boundary more quickly than those of lower mobility. These potentials are difficult to reproduce, tend to be unstable, and are seldom known with any accuracy; so steps must be taken to minimise them. Using 4 Molar KCL as the inner filling solution has the advantage that the K+ and Cl- ions have nearly equal mobilities and hence form an equi-transferrent solution. Also, in the single junction electrodes, the electrolyte concentration is much higher than that of the sample solution thus ensuring that the major portion of the current is carried by these ions. A third factor in minimising the junction potential is the fact that there is a small but constant flow of electrolyte out from the electrode thus inhibiting any back-diffusion of sample ions - although this is less important with modern gel electrolytes.

As indicated above, all these problems are doubled when double junction reference electrodes are used and an additional problem arises in the case of the last three listed above (Sodium Chloride, Ammonium Sulphate, Magnesium Sulphate) because the filling solutions are not equi-transferrent and hence have a stronger tendency to form liquid junction potentials. It must be noted here that Nico2000 Ltd have recently introduced a novel Lithium Acetate reference electrode which overcomes most of these problems and can be used with all the ELIT range of ISEs. This is because it contains ions which are very nearly equi-tranferrent and which do not interfere with any of the commonly used ISEs.

It must be noted that the E0 factor in the Nernst equation is the sum of all the liquid junction potentials present in the system and any variation in this during analyses can be a major source of potential drift and error in measurements.

d) Combination Electrodes

The majority of pH electrodes are produced in the form of combination electrodes in which the reference system is housed in the same cylindrical body as the sensor head. This produces a simple, compact unit for immersing in the test solution and has the added advantage that the two cells are in close proximity (with the reference cell normally completely surrounding the sensor element) - thus minimising the effect of any stray electrostatic fields or any inhomogeneity in the test solution. The main disadvantage of this arrangement is the fact that it is the reference element which is the most likely to cause problems or fail, long before the ISE head does, but the whole unit has to be replaced when failure does occur.

In contrast to pH electrodes, some ISEs are produced as mono-electrodes for use with separate reference systems. One reason for this is because ISE membranes have a far lower impedance than pH sensors and are less susceptible to stray electrostatic fields. Thus it is not necessary to screen the sensor head by surrounding it with the reference system. More importantly, the membranes and internal construction of ISEs are generally far more expensive than pH sensors and it is much more cost-effective to have separate units in which the reference system can be replaced independently from the ISE.

e) Multiple Electrode Heads: Separable Combinations.

A new concept for combination electrodes has recently been introduced. Both the ISEs and the reference electrodes are made in the form of 8mm diameter tubes fitted with a gold plated plug-in connector. These can be inserted separately into special multiple electrode heads which are fitted with the cables and connectors for attaching to the measuring system. The rigid plastic head ensures that the ISE and reference system remain firmly linked together at a regular distance apart during operation, but either one can easily be replaced in the event of failure or need to change the analysis. Moreover, the replacement electrodes are relatively inexpensive compared to conventional electrodes because they do not incorporate the expensive low-noise cables.


Related Solutions

Describe at least 3 types of aggregate healthcare reports that you believe would be commonly used...
Describe at least 3 types of aggregate healthcare reports that you believe would be commonly used by healthcare executives / administrators in a hospital or other healthcare setting, including public health agencies (only need to choose 1 healthcare setting).
Describe the two tools commonly used to evaluate an organizations external and internal situation.
Describe the two tools commonly used to evaluate an organizations external and internal situation.
a) Briefly describe TWO (2) types of stepper motor that are commonly used. b) A linear...
a) Briefly describe TWO (2) types of stepper motor that are commonly used. b) A linear induction motor drives a conveyor belt at speed of 50 km/h with a slip of 20% at 50 Hz and developed a thrust of 300 N. Determine: i) Pole pitch of the motor. ii) Power developed by the motor. iii) Copper loss in the secondary side. iv) The new value of synchronous speed if pole pitch is increase to 0.5 m.
Identify the different types of plans commonly used in the business context. Using examples from any...
Identify the different types of plans commonly used in the business context. Using examples from any published news sources, when each type of plan be most effective to use? 250 or less
The two most commonly used methods of capital budgeting analysis are the A. internal rate of...
The two most commonly used methods of capital budgeting analysis are the A. internal rate of return and net present value methods. B. net present value and payback methods. C. profitability index and the internal rate of return methods. D. net present value and discounted payback methods. E. average accounting return and discounted payback methods.
Which of the following statistical tests is commonly used to test a hypothesis? a- chi-square test...
Which of the following statistical tests is commonly used to test a hypothesis? a- chi-square test b-Student t-test c- Z-test d- All of the above
3.Phosphate-buffered saline (PBS) is a buffer solution commonly used in biological research. Most formulations contain a...
3.Phosphate-buffered saline (PBS) is a buffer solution commonly used in biological research. Most formulations contain a mixture of KCl and NaCl buffered to a pH of 7.40 using phosphate-containing compounds. a.The following reagents are available in your lab. Choose the correct reagents and calculate the volumes necessary to prepare exactly 5.0 L of a phosphate buffer with a pH of 7.40.                                                0.100 M H3PO4 0.150 M Na2HPO4                                                0.200 M KH2PO4 0.120 M Na3PO4 b.Commercial PBS solutions are typically prepared...
The most commonly used test for HIV has a sensitivity of 0.997 and a specificity of...
The most commonly used test for HIV has a sensitivity of 0.997 and a specificity of 0.985. In other words, a person infected with HIV will test positive for the virus 99.7% of the time while a person NOT infected with HIV will test NEGATIVE for the virus 98.5% of the time. Research current rates of infection for the indicated population in order to answer the following questions. If a US randomly selected US resident is tested for HIV and...
“Internal Rate of Return and Net Present Value are two of the most commonly used capital...
“Internal Rate of Return and Net Present Value are two of the most commonly used capital budgeting techniques”. In this light, answer the following questions: a) Discuss thoroughly with examples, what is meant by “Capital budgeting” b) Evaluate these two techniques considering their definitions, decision criteria and conflicting rankings. [10 Marks] c) If you were a financial manager, discuss thoroughly with examples which technique would you prefer to apply in making your capital budgeting decisions?
List the components of the framework commonly used in corporations when analyzing their internal control systems:...
List the components of the framework commonly used in corporations when analyzing their internal control systems: 1.) 2.) 3.) 4.) 5.) List the five common principles of Internal Control and give an explanation and example of each 1.) 2.) 3.) 4.) 5.) What are the two basic reasons that a company’s cash records do not balance to the bank statement? 1.) 2.)
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT