Absorption Spectrum - The curve representing
the light absorbed at each wavelength by pigment is called
absorption spectrum.
Procedure-
Sample
preparation-
- 1. Prepare an extract of chlorophyll by weighing 0.3 grams of
your leaf
sample. Tear the leaf section into small pieces and place the
tissue in a mortar
for grinding.
2. Add 10 mL of ethanol to the sample and grind with the pestle
until only tiny
specks of leaf tissue remain. Add an additional 20 mL of ethanol to
the
mortar.
3. Filter this liquid into a beaker through a funnel lined with a
coffee filter.
4. Dispose of the filter in the designated trash bin.
Calibration- The extract
must fall into a specified range to provide a reading with the
spectrophotometer. A sample that is too concentrated will not
provide a valid measurement. To assure the sample falls within a
readable concentration range, follow the steps provided.
- 1. Set the spectrophotometer wavelength to 640 nm.
2. Push the A/T/C button to select % transmittance. The digital
readout will show
%T.
3. Label one of the 3 mL syringes ethanol. Label the second syringe
leaf extract.
4. Dispense 3 mL of ethanol into a clean glass cuvette. (Wipe off
finger prints
with a Kimwipe.) Important: This cuvette is a “blank” that will be
used
throughout this lab to calibrate the spectrophometer.
5. Insert the blank in the cell holder and close the door.
6. Press the 0 ABS/100%T key to set the blank to 100%
transmittance.
7. The blank provides a standard for reading. Remove the blank and
proceed
with testing the extract by following the same procedure.
8. Dispense 3 mL of the extract into a clean glass cuvette. (Wipe
off all
fingerprints with a Kimwipe.)
9. Insert the cuvette in the cell holder and close the door.
10. A reading of 30%-40% transmittance is best. If %T is greater
than 40%, you
may skip the dilution step. If it shows less than 30%
transmittance, dilution of
the sample is required.
11. To dilute the filtered extract, add 0.5 mL of ethanol (using
the clean ethanol
syringe). Tightly hold a finger over the cuvette and invert to mix.
Re-test the percent transmittance. Continue with 0.5 mL increments
of ethanol until the
reading is greater than 30% transmittance.
Testing plant extract -
- 1.note the readings range from 400 nm – 700 nm (with increments
of 20nm). This is the range for testing.
2. Set the wavelength on the spectrophotometer to 400 nm.
3. Set the mode to “absorbance” by pressing the A/T/C button until
an A
appears at the right of the digital display.
4. Insert the blank (clean with a Kimwipe) in the cell holder and
close the door.
5. Press the 0 ABS/100%T key to set the blank to 0
absorbance.
6. Remove the blank and insert the clean cuvette containing the
chlorophyll
extract.
7. Record the digital readout for the absorbance in the data
table.
8. Continue to increase the wavelength by 20 nm each time and
repeat steps
4-7. Remember to use the blank for each new wavelength
setting!!
9. Graph the data for absorption spectrum.