Question

In: Chemistry

Electroanalysis a) List four technique that can be applied to improve the speed,sensivity and detection limits...

Electroanalysis

a) List four technique that can be applied to improve the speed,sensivity and detection limits of polarography?

b) Describe how an electrogravimetric analysis of a copper solution might be performed?

Solutions

Expert Solution

a. Using tast polarographic technique-current is only measured at the end of each drop lifetime-improvement the detection limit as capacitive component is supressed.

Differential pulse polarographic technique- current is measured before the beginig and before the end of short potential pulses-improve the detection limit

elecrode is polarised by means of a double voltage step to improve the signal to noise ratio to give more sensitivity.

By rapid scanning polarographic technique speed can be controlled.

b. As attached file

ELECTROGRAVIMETRIC ANALYSIS OF COPPER

INTRODUCTION

Most copper-containing compounds can be electrolyzed in acidic solutions, with quantitative deposition of Copper (Cu) at the cathode.. When a Platinum (Pt) anode and cathode are placed in solution of Cu2+, and an electric current is passed through, a deposit of copper metal will occur on the cathode and Oxygen (O2) will be liberated from the anode. Deposition occurs with the following reaction:

            Cathode:       2Cu2+   +   2e-        ↔    Cu(s)

           

            Anode:                            H2O    ↔    ½ O2(g)   + 2H+        +   2e-   

                                    ------------------------------------------------------------

            Net reaction:           H2O     +   Cu2+    ↔    Cu(s)       + ½ O2(g)   +   2H+             (17-1)1

In electrogravimetric analysis, the electrode is carefully weighed before and after deposition and the increase in mass tells how much analyte was deposited. There are three ways to observe completeness of the deposition. The first is to observe the disappearance of the color of the solution. This is seen because the colored species is the Cu2+, which is a light blue in color. The second observation is to expose most of the surface of the cathode to the solution during electrolysis by then adding water to the solution and seeing if more copper is deposited on the previously uncovered Pt. This is the method used in this lab. Alternately, remove a drop of solution and perform a qualitative test for copper. The samples must be free of chloride and nitrous acid.

SAMPLE PREPARATION

1.        Obtain unknown sample

2.        Weigh one gram of unknown sample on an analytical balance and transfer completely to a 250mL beaker.

3.        Add about x mL of 8M HNO3 and heat on a hot plate to dissolve the sample into solution.

Warning: Great care should be taken in the handling of nitric acid. Nitric acid can cause potentially severe chemical burns that result in the skin turning yellow and flaking off.  

4.        When the solid has dissolved, add distilled water for a final volume of 150mL.

CLEAN THE ELECTRODE

1.        Handle the Pt gauze with a kimwipe.

Caution: When handling the Pt gauze cathode, only touch the thick stem and not the gauze. This could result in contamination or transfer of fingerprints or damaging the cathode.

2.        Immerse the electrode in hot 8M HNO3 to remove previous deposits.

3.        Rinse with distilled water.

4.        Rinse with Ethanol.

5.        Dry in 105°C oven for 5 min, cool in desiccator for 5min.

6.        Weigh accurately using analytical balance and record weight.

Note:   If the electrode contains grease or any material that does not come off in nitric acid, it can be heated to red heat over a burner after the treatment above4.

SET UP THE ELECTROLYSIS SYSTEM

1.        Set up the electrolysis system as seen in figure 2.

2.        Position the cathode so that the top 5mm are above the liquid level after magnetic stirring     has begun.

3.        Adjust current to 2 A, which should require 3-4 V.

Warning:   The required current needed is enough to kill an adult human. Do not touch the connector pins while the direct-power supply is in operation, at the risk of severe electric shock.

4.        Add some distilled water when the blue color of Cu(II) has almost disappeared so that new Pt surface is exposed.

Note:   If no further deposition of Cu occurs on fresh surface in 15 min, the electrolysis is complete. If deposition is observed, continue electrolysis and test the reaction for completeness again.

DETERMINE AMOUNT OF COPPER

1.        Without turning off the power, lower the beaker while washing the electrode with distilled water.

2.        Turn the current off.

Caution:   If current is disconnected before removing the cathode from the liquid and rinsing off the acid, some Cu could dissolve and electrolysis will have to be repeated.  

3.        Wash cathode gently with water and alcohol.

4.        Dry at 105°C for 3 min, cool in a desiccator for 5 min, and weigh.

5.        Record final mass.

CONCLUSION

Determine the amount of copper present in your ore sample. It will contain between 5 and 20 percent. This experiment will be performed twice to increase the accuracy of the readings and obtain a smaller margin of error.

  


Related Solutions

Explain how a qualitative anatomical analysis can be applied to improve training technique. Why would a...
Explain how a qualitative anatomical analysis can be applied to improve training technique. Why would a clinician or practitioner want to do this and how could it help?
How is probability theory applied in fraud detection?
How is probability theory applied in fraud detection?
Describe a tool or technique which can be used to analyse and improve the performance of...
Describe a tool or technique which can be used to analyse and improve the performance of manufacturing systems e.g. Rank-order clustering, Ishikawa diagrams, value stream mapping etc. and explain its basic principle, how it would be applied, the potential benefits and inhibitors to implementation.         
Police plan to enforce speed limits by using radar traps at four different locations within the...
Police plan to enforce speed limits by using radar traps at four different locations within the city limits. A person who is speeding on his/her way to work has probabilities of 0.2, 0.1, 0.5, and 0.2, respectively, of passing through these locations (location 1, location 2, location 3, and location 4, respectively). The radar traps at each of the locations (location 1, location 2, location 3, and location 4, respectively) will be operated 40%, 30%, 20%, and 30% of the...
list FOUR VERY SPECIFIC dietary and or health realted goals one can implement to improve overall...
list FOUR VERY SPECIFIC dietary and or health realted goals one can implement to improve overall health status and explain how it does.
List and explain four ways in which hypothesis testing using big data can improve competitive advantage...
List and explain four ways in which hypothesis testing using big data can improve competitive advantage and decision making for businesses.
list the detection methods one can use to observe analyte spots that are invisible to the...
list the detection methods one can use to observe analyte spots that are invisible to the naked eye on a thin layer chromatography plate (5 methods)
Describe in some detail a western blotting and a ligand blotting technique for the detection and...
Describe in some detail a western blotting and a ligand blotting technique for the detection and comparison of relative levels of a cellular protein (e.g. receptor) in two different cell lines.
1. How can quality improvement techniques be applied to improve the effectiveness and efficiency of health...
1. How can quality improvement techniques be applied to improve the effectiveness and efficiency of health organizations? 2. Define federalism. Give an example of federalism.
1. Discuss the applicatoin of distillation as a separation and purificatoin technique. Are there limits due...
1. Discuss the applicatoin of distillation as a separation and purificatoin technique. Are there limits due to sample size, initial purity and final purity desired? 2. Discuss the applicatoin of gas chromatogrpahy as a separations and identification technique. Are there limits due to purity, sample size, phase of sample?
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT