Question

In: Chemistry

Calibration of a glass electrode gave a reading of 139.7 mV with saturated potassium hydrogen tartrate...

Calibration of a glass electrode gave a reading of 139.7 mV with saturated potassium hydrogen tartrate buffer standard (pH = 3.557) and a reading of -92.2 mV with 0.08 m HEPES, 0.08 m NaHEPES, 0.08 m NaCl buffer standard (pH = 7.454), both measured at 30°C.

(a) What is the observed slope (mV/pH unit) of the calibration curve?

(b) What is the theoretical slope at 30°C?

(c) What is β, the electromotive efficiency to three significant figures?? Open the hint panel below for the definition of electromotive efficiency.

(d) What is the pH of an unknown that gives a reading of 46.9 mV with this electrode at 30°C?

Solutions

Expert Solution

a.

Slope S = dE/dpH = (139.7-(-92.2))/(3.557-7.454) = - 231.9/3.897= - 59.51 mV/pH unit

b. The theoretical slope is

          -2.303RT/nF

    The known value at 298 K is - 0.05916 V (see Nernst equation)

     The value at 303 K is

        -0.05916 V x303K/298K = -0.06015 V = -60.15 mV/pH unit

c.

59.51 mV/pH / 60.15 mV/pH = 0.9893

d.

pHx = pH + (Ex-E)/S      

                                          (pH and E, any known pair, look also at a., same equation)

                = 7.454 - (46.9 + 92.2)/ 59.51

                = 7.454 – 2.337 = 5.117


Related Solutions

A photometer with a linear response to radiation gave a reading of 690 mV with a...
A photometer with a linear response to radiation gave a reading of 690 mV with a blank in the light path and 101 mV when the blank was replaced by an absorbing solution. Calculate the transmittance and absorbance of the absorbing solution. Transmittance = Absorbance = the expected transmittance if the concentration of absorber is one half that of the original solution. Transmittance = the transmittance to be expected if the light path through the original solution is doubled. Transmittance...
What is the difference between the Normal Hydrogen Electrode (NHE) and the Reversible Hydrogen Electrode (RHE)?...
What is the difference between the Normal Hydrogen Electrode (NHE) and the Reversible Hydrogen Electrode (RHE)? - Also, if a redox happens at the NHE of 0.2 V, please calculate the potential for the RHE at pH = 1.
What is the difference between the Normal Hydrogen Electrode (NHE) and the Reversible Hydrogen Electrode (RHE)?...
What is the difference between the Normal Hydrogen Electrode (NHE) and the Reversible Hydrogen Electrode (RHE)? - Also, if a redox happens at the NHE of 0.2 V, please calculate the potential for the RHE at pH = 1.
Consider a glass electrode or [pH electrode (combination electrode). a. Identify the reference part. b. Identify...
Consider a glass electrode or [pH electrode (combination electrode). a. Identify the reference part. b. Identify the indicator part. c. Explain how the potential of the electrode is determined. d. Describe a two-point calibration of the electrode. e. State any five of the eight possible errors that might occur in pH measurements with the electrode.
A galvanic cell consists of a standard hydrogen electrode and a copper electrode. Suppose that the...
A galvanic cell consists of a standard hydrogen electrode and a copper electrode. Suppose that the copper electrode is immersed in a solution that is 0.1 M NaOH and that it is saturated with Cu(OH)2. Find the cell potential?
Ag/AgCl electrode, saturated KCl used as electrolyte. In Hg2Cl2 (reference electrode), Hg/Hg2Cl2 used as electrode, 1M...
Ag/AgCl electrode, saturated KCl used as electrolyte. In Hg2Cl2 (reference electrode), Hg/Hg2Cl2 used as electrode, 1M saturated KCl used as electrolyte. Their standard half-cell potentials are E0S=+0.223V and E0R=+0.268V, respectively. They are combined to form a full cell for concentration sensing, i.e., as a concentration cell. 1) What type of electrode will it be (potenometry or amperometry)? 2) What is the anticipated output value from the sensor if the sampled solution has 1M concentration? (T=250C). 3) If the output value...
1. a. How is electrode potential related to electrochemical equilibrium? b. When the standard hydrogen electrode...
1. a. How is electrode potential related to electrochemical equilibrium? b. When the standard hydrogen electrode is connected to a half-cell containing magnesium metal immersed in 1.0 M MgSO4,, which electrode is the anode? Why? c. When the standard hydrogen electrode is connected to a half-cell containing an inert electrode in contact with liquid bromine and with 1.0M KBr, which electrode is the anode? Why?
What would the potential of a standard hydrogen (S.H.E.) electrode be if it was under the...
What would the potential of a standard hydrogen (S.H.E.) electrode be if it was under the following conditions? [H+]=0.59M   PH2 =3.0 atm T=298 K
Analyte in an unknown gave a signal of 25.0 mV. When 1.00 mL of 0.065 M...
Analyte in an unknown gave a signal of 25.0 mV. When 1.00 mL of 0.065 M standard addition reagent was added to 10.00 mL of the unknown sample, the signal increased to 35.0 mV. What is the analyte concentration of the original unknown?
The standard electrode use to reference all electrode potentials to is? A.chlorine/chloride B.Cu2+/copper C.proton/hydrogen D.Zn2+/zinc
The standard electrode use to reference all electrode potentials to is? A.chlorine/chloride B.Cu2+/copper C.proton/hydrogen D.Zn2+/zinc
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT